Hello,
we are trying to certify a Class I medical device. The device is internally powered, i.e., it contains a Li-ion battery which supplies power during usage.
We are struggling to understand how to treat the associated charger, which is an external component which converts mains to a 19V. I know this question has been asked multiple times, but even the answers are very application-specific and sometimes contradictory.
The charger is shipped toghether with our core product and supplied off-the-shelf, and by itself is NOT 60601 compliant; however, it has all the standard certification for a quality non-medical charger and comes from a reputable manufacturer. It is a traditional charger that goes into mains, has a transformer, and then another cable which supplies the voltage to the DC port of our device.
The charger is not meant to be connected while the device is used, as functionality is blocked; however, it is possible (foreseeable misuse) for the patient to wear the device while keeping the charger connected, despite no functionality is present and clear indications to avoid doing so have been provided.
We are evaluating the following options:
Note in Subclause 7.9.2.14 is even more explicit, but does not mention ACCESSORIES at all:
To add a bit more of entropy, we have performed some external lab tests in which the charger was shipped with our product; some tests (i.e., leakage currents) have been performed taking into account the charger. Some others (i.e., IP22, drop tests, ...) did NOT take into account the charger, nor the device+charger (connected) system.
we are trying to certify a Class I medical device. The device is internally powered, i.e., it contains a Li-ion battery which supplies power during usage.
We are struggling to understand how to treat the associated charger, which is an external component which converts mains to a 19V. I know this question has been asked multiple times, but even the answers are very application-specific and sometimes contradictory.
The charger is shipped toghether with our core product and supplied off-the-shelf, and by itself is NOT 60601 compliant; however, it has all the standard certification for a quality non-medical charger and comes from a reputable manufacturer. It is a traditional charger that goes into mains, has a transformer, and then another cable which supplies the voltage to the DC port of our device.
The charger is not meant to be connected while the device is used, as functionality is blocked; however, it is possible (foreseeable misuse) for the patient to wear the device while keeping the charger connected, despite no functionality is present and clear indications to avoid doing so have been provided.
We are evaluating the following options:
- Treat the charger of part of our ME EQUIPMENT.
- Which are the implications? Do we need to re-label and claim responsibility for the charger? For instance, do we need to include the materials of the charger in our biocompatibility assessment?
- Do we need to perform tests for MAINS PART connection in this case?
- Do we need to account for the charger rating when supplying the rating of our device?
- Treat the charger as part of an ME SYSTEM made of our ME EQUIPMENT device and the charger.
- Does this opens us to any additional testing arising from 60601-1 Clause 16?
- Does the carger classify as a power supply, even if the power is sourced from the battery during usage?
- Treat the charger as an ACCESSORY. this is what was suggested from our test lab; however, we are not convinced of this narrative as the charger only serves for recharging, i.e., does not help the device achieve its INDENDED (medical) USE.
So, it suggests ACCESSORIES need to be considered as such only when the medical (INTENDED) purpose is affected by the employment of the accessory; however, in NOTE 1 of 3.63 MEDICAL ELECTRICAL EQUIPMENT definition:ACCESSORY
additional part for use with equipment in order to:
– achieve the INTENDED USE,
– adapt it to some special use,
– facilitate its use,
– enhance its performance, or
– enable its functions to be integrated with those of other equipment
So, it refers to NORMAL USE, i.e., also includes maintenance, transport, ... and is in general better applicable to a charger which does not have any medical purpose nor certification.NOTE 1 ME EQUIPMENT includes those ACCESSORIES as defined by the MANUFACTURER that are necessary to enable the NORMAL USE of the ME EQUIPMENT.
Note in Subclause 7.9.2.14 is even more explicit, but does not mention ACCESSORIES at all:
NOTE What was referred to in the first and second editions of this standard as a “specified power supply” is considered either as another part of the same ME EQUIPMENT or as other equipment in an ME SYSTEM. Similarly, a battery charger is considered either as part of the ME EQUIPMENT or as other equipment in an ME SYSTEM.
To add a bit more of entropy, we have performed some external lab tests in which the charger was shipped with our product; some tests (i.e., leakage currents) have been performed taking into account the charger. Some others (i.e., IP22, drop tests, ...) did NOT take into account the charger, nor the device+charger (connected) system.