Hi all
I was hoping someone could provide some input on the different timing requirements for alarm signals vs alarm pointers.
We have some software that was using the urgency encoded alarm signals (section 1.1) from this site: th.id.au/alarms/. After some external validation testing, it was reported that the timing between pulses in the high priority alarm with not in compliance with table 3 in clause 6.3.3.1.
After some investigation I uncovered the alarm pointers provided in annex G of amendment 2. However, table G.1 and G.2 have a different set of timing requirements for the pointers vs the alarms. This is confusing!