IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure

A

ale_pb

Hello

My question is:

Let's say that I have a CLASS I EQUIPMENT, with a plastic enclosure and no accessible metal parts.
Reading the 60601-1 it seems to me that, if I don't have any accessible metal parts, the ground bond test is not necessary. Am I correct?

Best Regards
 

Pads38

Moderator
Re: Ground bond test - Plastic enclosure

The ground bond test verifies the impedance of the protective earth. If you have a protective earth (defined term in 60601-1) then you need the test.

For a type approval test this could be done 'with the lid off' but it is also one of the tests that a hospital service department would expect to do and you run the risk of problems if you don't make it easy for them. (Things like trying to use 25A test current on a point that goes via a small circuit board trace).

I have found that adding an equipotential point on the rear panel gives a nice easy place to test without being difficult or expensive to fit.
 

Marcelo

Inactive Registered Visitor
Let's say that I have a CLASS I EQUIPMENT, with a plastic enclosure and no accessible metal parts.
Reading the 60601-1 it seems to me that, if I don't have any accessible metal parts, the ground bond test is not necessary. Am I correct?

For your equipment to be class I, it has to have accesible or internal parts that are protectively earthed. As yours do not have accessible protectively earthed parts, i'm assuming your equipment has internal parts which are protectively earthed.

The ground bound test verify that protectively earthed parts, accessible or internal, have the right impedance. So, assuming that your have those internal parts, the test has to be done in those parts.

If your equipment do not have protectively earthed parts, it cannot be class I.
 

Peter Selvey

Leader
Super Moderator
Under IEC 60601-1:1988/A2:1995 this is poorly written part of the standard so easy to overlook.

Clause 18g covers only accessible protectively earthed metal parts, with the typical 25A / 0.1ohm test.

Clause 18f covers "other" protectively earthed metal parts and then goes on to describe an alternate test. The way it is written means it is often mistakenly considered not applicable if the alternate test is not applied, and the IECEE CB scheme protocol has promoted this mistake (many test engineers read only the protocol, not the standard). As such, the test of internal protectively earthed parts is frequently overlooked.

But the correct interpretation is that for internal parts, either the test of 18g (i.e. 0.1ohm / 25A) or the alternate test applies. Either way, internal protectively earthed parts must be tested with the result recorded under Clause 18f.

The 3rd edition, IEC 60601-1:2005 has removed the word "accessible" from the applicable clause, so the problem no longer exists.
 
Thread starter Similar threads Forum Replies Date
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Ground stud requirements on our medical device in IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Sample qualification to be IEC 60601-1 series tested IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
G History of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Safety testing IEC 60601 series Medical Device and FDA Regulations and Standards News 2
F Battery overload in IEC 60601-1 vs 62133-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Mandatory tests for IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
P IEC 60601-1-2 RF wireless technology IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-2-8 cl.201.10.1.2.101 and cl.201.10.1.2.104 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G Risk Management for IEC 60601-1 and IEC 60601-1-2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 15
J IEC 60601 Applied Part - Probes intended to be covered by sheaths IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601-1-11 Requirements related to Storage and Distribution IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
shimonv IEC 60601-1-2:2014+A1:2020 (Ed. 4.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G IEC 60601 Compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R IEC 60601-2-57 light source classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
L Top opening of enclsoure against fire in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EN 45502 vs IEC 60601-1-8 Other Medical Device Related Standards 0
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9

Similar threads

Top Bottom