Hello to all,
we are legal manufacturer of medical body electrodes (ECG, Neutral Electrodes, Defibrillation Electrodes). We are preparing a biological evaluation according ISO 10993-1:2018 for our ECG Electrodes. The ECG electrodes are classified as follows: class I in EU according REG (EU) 2017/745 (MDR), class II in USA according 21CFR870.2360. The ECG electrodes have been marketed since many years.
They were approved with former ISO 10993-1:2009.
In the case of the ECG electrodes, the nature of body contact is indicated with Surface device intended to be used on intact, uninjured skin and the
contact duration is indicated with "limited (<24h)" and prolonged (24 h up to 30 d).
Consequently, the ISO 10993-1:2009 standard requires the assessment of
- Cytotoxicity
- Irritation or intracutaneous reactivity
- Sensitization
Now, our consultant who is helping us with the biological evaluation claims, that we have to perform chemical characterization for materials according ISO 10993-18. We recognized, that it is demanded in ISO 10993-1:2018. But we argue with clause 4.11 from ISO 10993-1:2018, which indicates:
[...]
4.11 This document shall not be used to mandate re-testing of historical products assessed previously using the appropriate edition of this document at the time of the assessment. Nevertheless, compliance to this new edition shall be shown, by providing a justification for omission of further testing. Where recommendations for endpoint assessment per Annex A are different from prior published versions of this document, a history of safe clinical use can be used to document why additional testing on a commercially-marketed medical device is not needed. However, if any of the changes described in Clause 4.9 occur, an evaluation of the biologic risks related to the change shall be performed using the current version of this standard.
[...]
We think, that we do not have to perform new tests (chemical characterization), as we only have historical devices, marketed for a long time. We just have to write this in the biological evaluation and reference to the clinical evaluation report, as well.
But our consultant does not accept this argumentation and insists on chemical characterization testing. By the way, the consultant does not benefit from new testings.
I would appreciate your opinions. What do you think about omitting chemical characterization by using clause 4.11 from ISO 10993-1:2018.
Thank you for your help.
we are legal manufacturer of medical body electrodes (ECG, Neutral Electrodes, Defibrillation Electrodes). We are preparing a biological evaluation according ISO 10993-1:2018 for our ECG Electrodes. The ECG electrodes are classified as follows: class I in EU according REG (EU) 2017/745 (MDR), class II in USA according 21CFR870.2360. The ECG electrodes have been marketed since many years.
They were approved with former ISO 10993-1:2009.
In the case of the ECG electrodes, the nature of body contact is indicated with Surface device intended to be used on intact, uninjured skin and the
contact duration is indicated with "limited (<24h)" and prolonged (24 h up to 30 d).
Consequently, the ISO 10993-1:2009 standard requires the assessment of
- Cytotoxicity
- Irritation or intracutaneous reactivity
- Sensitization
Now, our consultant who is helping us with the biological evaluation claims, that we have to perform chemical characterization for materials according ISO 10993-18. We recognized, that it is demanded in ISO 10993-1:2018. But we argue with clause 4.11 from ISO 10993-1:2018, which indicates:
[...]
4.11 This document shall not be used to mandate re-testing of historical products assessed previously using the appropriate edition of this document at the time of the assessment. Nevertheless, compliance to this new edition shall be shown, by providing a justification for omission of further testing. Where recommendations for endpoint assessment per Annex A are different from prior published versions of this document, a history of safe clinical use can be used to document why additional testing on a commercially-marketed medical device is not needed. However, if any of the changes described in Clause 4.9 occur, an evaluation of the biologic risks related to the change shall be performed using the current version of this standard.
[...]
We think, that we do not have to perform new tests (chemical characterization), as we only have historical devices, marketed for a long time. We just have to write this in the biological evaluation and reference to the clinical evaluation report, as well.
But our consultant does not accept this argumentation and insists on chemical characterization testing. By the way, the consultant does not benefit from new testings.
I would appreciate your opinions. What do you think about omitting chemical characterization by using clause 4.11 from ISO 10993-1:2018.
Thank you for your help.