Hi,

We have both Gamma and ETO Sterilized products. Wanted to know if we need to select one product each quarter for both ETO and Gamma sterilized products or we can pick one from either Gamma or ETO product to perform quarterly monitoring.

Thanks,
Bala
 
Elsmar Forum Sponsor

planB

Super Moderator
#2
Bala,

in case your gamma and EO sterilized products represent each other in terms of bioburden and factors having an impact on bioburden, you could theoretically do so. You may want to keep in mind, that validation of irradiation sterilization typically follows a bioburden-based approach (e.g. VDmax25), while EO sterilization predominantly follows a BI half-cycle overkill approach. Thus, the level of required bioburden control might be different, depending on your sterilization modality.

HTH,
 

planB

Super Moderator
#5
Bioburden should be done for each product. You can't extrapolate bioburden results from one product to another product.
ISO 11737-1:2018 allows definition of bioburden families - quote:
5.1.2 If product(s) are grouped in a product family for the purpose of the determination of bioburden, the rationale for inclusion of a product within a product family shall be recorded. The rationale shall include criteria to ensure that bioburden determined for a product selected from the product family is representative for the whole product family
 

indubioush

Quite Involved in Discussions
#6
ISO 11737-1:2018 allows definition of bioburden families - quote:
Ah, yes. Thanks for bringing this up. I was making the assumption that the two products are completely different because they have different sterilization methods, but this may not be the case. Perhaps the OP can fill us in on whether these are products within the same family.
 
Thread starter Similar threads Forum Replies Date
E Routine production tests - dielectric voltage withstand test IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
T Nonconforming product at customer detected by a routine inspection by field service Nonconformance and Corrective Action 9
M Routine testing of medical electrical systems - What specific electrical safety tests should be performed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Referencing of a standard's year in routine regulatory compliance documents Other Medical Device Related Standards 3
M Medical Device News FDA - Eliminating Routine FDA news 14-09-18 - Re-Review of Third Party 510(k) Reviews Other US Medical Device Regulations 0
S Routine Testing for Class I Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 24
N ISO 11135 Monitoring Routine ETO Sterilisation Parameters Other Medical Device Related Standards 3
C Dose Mapping - How to relate back to Routine Monitoring Position ISO 13485:2016 - Medical Device Quality Management Systems 3
B Number of Biological Indicators for Routine Monitoring of ETO Sterilization Cycle Other Medical Device Regulations World-Wide 8
J FDA Routine Inspection Pre-Anouncement - What will be checked? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M Identifying Routine and Non-Routine Activities: OSHA OHSAS 18001 Occupational Health & Safety Management Standards 6
S How you define 'Normal & Abnormal' Conditions and 'Routine & Non-routine' Activities ISO 14001:2015 Specific Discussions 8
J Routine test for MD finished product IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
RoxaneB Environmental Objectives - Do they need to be linked into with Routine Indicators? Miscellaneous Environmental Standards and EMS Related Discussions 6
RoxaneB Search for another name for "Standardization Process" or "Routine Management" Quality Tools, Improvement and Analysis 4
D Line clearance - Routine components in the prep room for universal use ISO 13485:2016 - Medical Device Quality Management Systems 8
R Inspection and Detection - Kit to production - Need an inspection routine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
M Bioburden Frequency (sterile MD) Other Medical Device Related Standards 2
K When is Bioburden Testing Required? Other Medical Device Related Standards 8
A ISO 11135:2014, B.1.4, BI resistance x product bioburden ISO 13485:2016 - Medical Device Quality Management Systems 6
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
B Radiation Sterilization Dose Audit: Bioburden frequency Other Medical Device Related Standards 1
F Bioburden Testing requirements for Non-Sterile Dental Products EU Medical Device Regulations 6
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
K Bioburden 0 cfu for Dose Setting using the VDmax25 Other Medical Device Related Standards 9
N Must Bioburden Sampling Procedures be Documented? Other Medical Device Related Standards 5
R Simulated product for bioburden testing Other ISO and International Standards and European Regulations 2
C What is the most appropriate Bioburden Test Method for Surgical Gloves ISO 13485:2016 - Medical Device Quality Management Systems 10
M Bioburden numbers and ECA gloves 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Bioburden testing for Class 10000 (Grade C) clean room ISO 13485:2016 - Medical Device Quality Management Systems 1
O Bioburden Testing - Non Sterile Product / Tapes - ISO 11737-1 Other ISO and International Standards and European Regulations 9
P Serial Testing of Bioburden for Medical Devices Other Medical Device Related Standards 2
M Biological Monitoring Program Environmental Bioburden Test Method Needed ISO 13485:2016 - Medical Device Quality Management Systems 1
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
P Frequency of Bioburden Testing EU Medical Device Regulations 3
B BioBurden and EndoToxin Questions Other Medical Device Related Standards 6
M Sterility Testing vs Bioburden Testing - Differences Qualification and Validation (including 21 CFR Part 11) 7
S Bioburden Monitoring Sample Size Increase ISO 13485:2016 - Medical Device Quality Management Systems 4
I Bioburden and Endotoxin (LAL) Testing proposal help 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
I Bioburden Count Question - Bioburden is above the alert limit >1 CFU/component avg. 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Bioburden Monitoring - 'Off the Shelf' Bioburden/Microbiological Monitoring Software ISO 13485:2016 - Medical Device Quality Management Systems 4
L GB15980-2003 Bioburden Count Trend on Operator Hands Other ISO and International Standards and European Regulations 1
0 Guideline/Method for Raw Material Roll Stock Bioburden Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
Q Bioburden Requirements on Device used on Unbroken Skin Other Medical Device and Orthopedic Related Topics 4
A Microbial Limit (Bioburden)on dry nonwoven that used for making Hand sanitizing wipes ISO 13485:2016 - Medical Device Quality Management Systems 1
D Bioburden and Product Labels - VD Max studies on all the BOMs Other Medical Device and Orthopedic Related Topics 9
H Determining a New Product Bioburden Limit ISO 13485:2016 - Medical Device Quality Management Systems 11

Similar threads

Top Bottom