Scope statement for CMDCAS

DEVigil

Involved In Discussions
Greetings, Fellow Covers!

I am having a dispute with my NB regarding our scope statement for ISO 13485:2003 (CMDCAS). They insist it must slavishly follow the template that Health Canada recommends in GD 207 for non-IVD devices; specifically that it must include a 'for the area of' statement.

Our current scope statement, which we hammered out in conjunction with the same NB two years ago, is "Design and development, Manufacture, installation, and service of medical device imaging and communication software systems." Essentially, our product offers Picture Archiving and Communication System (PACS)-type functionality.

Somehow, after two years of being acceptable, and after a surveillance (not renewal) audit, [FONT=&quot][/FONT]the NB claims that we need to add 'for the area of __' to the scope. Our software is not limited to a certain clinical scope - it can be used to display and communicate multiple kinds of medical images (e.g., RT, CT, PET scans, etc.) to facilitate whatever you would need to display/communicate them for. We do have a focus area in radiation oncology (which our NB apparently does not understand is distinct from radiology, which is the 'area of' their 'experts' want to add), but the functionality is not limited only to that.

In GD 207, Health Canada does not require adherence to the exact formula provided in their templates (which are, themselves, not exhaustive in terms of possible terms to include). They only 'strongly recommend' the format. Furthermore, there are illustrative examples - and those examples do not include all elements from the recommended templates. There is even an example (B2) with a hypothetical product that includes imaging-oriented functionality that does not have any 'for the area of' language at all. Health Canada plainly recognizes that a single template is not suitable for all use cases.

I've pointed this out, and asked for an explanation for why the current scope statement is suddenly no longer acceptable. Their response has amounted to "The template is mandatory, and we are experts and this is our opinion."

I do not want to limit my scope statement to a single 'area of' when that does not accurately represent the actual scope of the product's use. And I certainly do not accept their claim that such language is mandatory when the guidance document clearly states the template is only recommended, and its illustrative examples do not follow it slavishly. Help?

(P.S. - This entire issue was only communicated to me when I refused to pay the invoice for the recent surveillance audit, which took place over three weeks ago, because I had yet to receive the audit report....)
 

Ronen E

Problem Solver
Moderator
Hi,

You sound right, but in my opinion in this case you also have to be wise.

Normally, when one receives bad service the ultimate leverage is changing suppliers. Unfortunately, in the current context it doesn't seem to be a viable option. Seems you are stuck with them and will have to somehow plough through.

Geographically, is a face to face meeting feasible? When people are stuck together in a room they sometimes become more constructive and less arbitrary than over the phone.

How about a wording along the lines of "for the areas of radiology, radiation oncology and similar applications."? The word "similar" might be flexible enough to prevent anyone from being bluntly non-compliant, so that all stakeholders are happy. Perhaps you can find an even better phrasing, since you are exposed to the details. I know it might seem like trying to dodge the system, but you've actually already tried the factual / rational approach and it seems to be failing. Besides, the whole thing seems to be a non-essential feature so it boils down to satisfying some stubborn bureaucrat.

Just my :2cents:

Ronen.
 
S

Sarah Stec

Howdy,

So you're right that a guidance document isn't The Word. But Health Canada sometimes acts like it is, and everyone has to fall in line.

HC Registrars are audited against their adherence in applying the guidance documents. A couple years ago HC started to really enforce the certificate scope provisions because some high percentage (I think a bit more than half, but that's off the top of my head a couple years after the fact) of the certificate scopes didn't conform to the GD 207. So your HC Reg may be reacting to their own nonconformities.

Another way to think about it: ISO 13485:2003 under CMDCAS is a regulatory requirement for HC, and so the certificate and scope needs to respond to the regulations (unlike "vanilla" ISO 13485:2003 by itself). The CMDR requires that the manufacturer include the purposes and uses of the device in its device license application. HC wants conformity between the device license and the QMS scope. If the manufacturer's QMS produces devices for a purpose different than that for which the manufacturer is applying for a license, how can HC assess if the manufacturer's system can produce the device fit for its intended purpose, or know what the real intended purpose is?

If your product is specialized enough to only be used in one area anyway, then you may not need it. Also, I've seen CMDCAS certificates with 5 or 6 areas - you should be able to have as many areas as you're able to support.

I hope that this is helpful! :bigwave:
 

DEVigil

Involved In Discussions
If Health Canada's own examples slavishly followed the recommended language, I might be inclined to agree. But they do not. And specifically, they do not for products whose scope isn't tied to particular areas of practice. Ours falls into that kind of category - image processing, display, and communication is useful across a broad spectrum. The second I list one, or even a group, I close off others.

The CMDR itself says blessed little about the QMS certificate, other than a discussion of its period of validity and what happens if it expires or is cancelled. Sections 32 and 33 are more about registrars. Presumably, that is why GD 207 exists. And as previously noted, GD 207 itself doesn't back the NB's position.
 

DannyK

Trusted Information Resource
DEVigil- I am a CMDCAS auditor and I would not have an issue with your scope.

From what I remember, the scope is manufacturer's responsibility and it is not the certification body's responsibility to make changes to the scope.

I would tell the certification body that if Health Canada has an issue with the scope, you will deal with Health Canada. Let Health Canada reject it.

If they refuse, contact Health Canada and ask them.
 

somashekar

Leader
Admin
A medical device with the same intended use can be operated under various environments, and each such use demands specific requirements being met by that device. So in order to get the clarity it is advised that the scope has "for the area of" clearly stated.
Example .... An oximeter can be for simple use in home area., in clinics, hospital and general ward area, in ICU or OT area, In Ambulance .... Etc.,
No do you see that the area of use differs. Now can you reflect upon your design inputs to see which specific area of use you designed for.
When you put in in the scope, you are defining a medical device usage with more clarity, which the HC expects. Please look at the scope from the point of view of such area, which clearly distinguishes your device for the group of standards it is considering.
 

DEVigil

Involved In Discussions
Somashekar,

I do not understand your reply. You seem to be discussing intended use, which is not the same as the scope statement for a certificate. We have a separate intended use statement.
 

somashekar

Leader
Admin
If your intended use statement details the area of use pretty much as I have detailed in my post, CMDCAS wants this said in the scope statement...
 

DEVigil

Involved In Discussions
Where a device is used/by whom it is physically used (as in, professional versus home) is a distinct issue from defining a clinical area of practice (e.g., radiation oncology). They are distinct topics with distinct definitions. I am only discussing the alleged need for the latter on a certificate, not the former.
 
Thread starter Similar threads Forum Replies Date
R How to define QMS certification scope statement? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
E Adding a "Scope Statement " to my Quality Policy ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
C Taking exceptions in the scope statement for registration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
P Scope Statement Content for OHSAS 18001 Occupational Health & Safety Management Standards 10
Jim Wynne Is this ISO 9001 Registration Scope Statement Too Vague? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
A Examples of a Scope Statement - CNC Machining, Sheet Metal Fabrication, Assembly ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
V ISO 27001 Sample Scope Statement example wanted IEC 27001 - Information Security Management Systems (ISMS) 5
J ISO9000:2000 Scope Statement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Z Design verification test sample size under the scope of first in human study submission in Europe Medical Device and FDA Regulations and Standards News 3
L Review of Scope of Mangement System ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Adding a Processing site to scope, shipping before certification? IATF 16949 - Automotive Quality Systems Standard 2
L AS9100 work for another site outside scope AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S Audit Duration Calculation - ISO9001 + EN9100 with different scope General Auditing Discussions 2
S Smart Scope Fixturing Quality Tools, Improvement and Analysis 3
B ISO 13485 Certificate Location / Scope ISO 13485:2016 - Medical Device Quality Management Systems 9
S 7.1.5.3.1 Internal Lab Scope IATF 16949 - Automotive Quality Systems Standard 15
Tagin Corporate HQ handles AR/AP - Outsourced or Out-of-scope? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D API Q1 clause (1.0) Scope Oil and Gas Industry Standards and Regulations 1
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
C ISO13485 scope as Legal Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 2
Ed Panek ISO 9001 Cert Scope Adjustment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
A Laboratory scope on the IATF IATF 16949 - Automotive Quality Systems Standard 5
I Restricting scope to one part number for one customer IATF 16949 - Automotive Quality Systems Standard 6
A ISO 9001- any advise please with transitioning from a paper based system to eQMS and multi site scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S F-108 Request for Expansion of Scope of Accreditation ISO 17025 related Discussions 2
S Scope of management review ISO 13485:2016 - Medical Device Quality Management Systems 5
D Do all products fall under scope of AS9100? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M PROBLEM IN SCOPE OF ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
N ISO 13485 Scope Reduction Other ISO and International Standards and European Regulations 2
M AS 9100 Scope for Repair Station with PMA approval AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M Intended Use vs Actual Use and Scope of Risk Management EU Medical Device Regulations 8
Stefan Mundt Split Scope = Separate Audits ??? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
lisap Design falling outside ISO scope Misc. Quality Assurance and Business Systems Related Topics 11
A Scope of ISO 13485 certification as legal manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
R ISO 13485 certification scope ISO 13485:2016 - Medical Device Quality Management Systems 5
Q Kap.4.3 Determining the scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B ERP software validation - risk assessment vs validation scope ISO 13485:2016 - Medical Device Quality Management Systems 11
G Why are 17025 scope angle measurements limited to 5 to 85 degrees? ISO 17025 related Discussions 2
3 Scope question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
G Internal Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 10
T How to handle out-of-scope activities. ISO 13485:2016 - Medical Device Quality Management Systems 7
Enghabashy Supply chain main policies ,scope, risk assessments & relavant KPI Supply Chain Security Management Systems 2
M Does the scope of ISO 9001:2015 applies to tenders, pricing and sales department of a medical devices distributor? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Can outsourced processes be excluded from QMS scope? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
D IATF 16949 M&TE Laboratory scope IATF 16949 - Automotive Quality Systems Standard 7
J Informal vs formal scope creep... managing non-medical devices through system processes ISO 13485:2016 - Medical Device Quality Management Systems 2
R "Medical devices" required in scope ISO 13485:2016 - Medical Device Quality Management Systems 2
R CB/Auditor Requiring a change in scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9

Similar threads

Top Bottom