I received a calibration certificate for my Hygrometer (which is a device that is used to monitor the humidity in lab room). The reported uncertainty is +-5%.
Going back to the test method, the required humidity in the lab room should not exceed 50+-5%.
The question is that if the hygrometer gave me a reading of 53%. Then according to the calibration certificate the reading may be 58% or 48%.
Which make me somehow not sure if I am complying with the test method requirement?!!!
should i accept the uncertainty in the calibration certificate or should i try to using different device with better readability or resolution to improve the uncertainty?
thanks