Hi All,
IEC 60601-1:2012 requires Mould Stress Relief testing to be performed on plastic enclosures and references Cl. 15.3.6 of 60601-1:2005. This clause appears to specify that the part be subjected to a minimum temp of 70C x 7 hrs. The primary question I have is, is it possible to avoid testing at 70C and instead perform the test at 50C -but for a longer duration (referencing an appropriate Aarhenius temp acceleration factor) and claim that the 50C test is equivalent to the 70C Mould Stress Relief Test? i.e. is the Aarhenius model valid for Mould Stress Relief Test?
As I understand it the test is a test of the design of the plastic part wrt the residual streses remaining after moulding. I note that only 1 sample is required for the test and that any damage sustained must not constitute an unacceptable risk wrt basic safety and essential performance of the final product.
My secondary question is; There is an alternative to doing this test as the standard states; [I]"Compliance is checked by inspection of the construction and available data where appropriate or by the following test." [/I]....and then it proceeds to list the 70C x7hrs test, (-so based on my reading of the standard anyway it would appear that if there is adequate data available it is not necessary to do the temp test.)
Can someone with previous experience of this scenario please advise what the "inspection" and "available data" referred to above might involve. And which approach most people follow ..the inspection/data approach or the temp test? I'm guessing that it would involve design validation activities, e.g. mold flow simulations, inspection & dimensional analysis for sink, warpage etc. verification of tool construction to incorporate tool design guidelines to minimise residual stresses, etc, etc, but I'm just guessing and I'd really value the experience of someone who has gone down this road already preferably with a medical device incorporating plastic parts.
In terms of background the plastics involved are a range of PP, HDPE, LDPE, TPE and POM parts. The product is a medical device and the Mould Stress Relief test will form part of the design verification testing.
All comments, suggestions and insights welcomed, (my primary question is the main one I want answered the secondary question is a "nice to know".)
Kind regards,
SR
IEC 60601-1:2012 requires Mould Stress Relief testing to be performed on plastic enclosures and references Cl. 15.3.6 of 60601-1:2005. This clause appears to specify that the part be subjected to a minimum temp of 70C x 7 hrs. The primary question I have is, is it possible to avoid testing at 70C and instead perform the test at 50C -but for a longer duration (referencing an appropriate Aarhenius temp acceleration factor) and claim that the 50C test is equivalent to the 70C Mould Stress Relief Test? i.e. is the Aarhenius model valid for Mould Stress Relief Test?
As I understand it the test is a test of the design of the plastic part wrt the residual streses remaining after moulding. I note that only 1 sample is required for the test and that any damage sustained must not constitute an unacceptable risk wrt basic safety and essential performance of the final product.
My secondary question is; There is an alternative to doing this test as the standard states; [I]"Compliance is checked by inspection of the construction and available data where appropriate or by the following test." [/I]....and then it proceeds to list the 70C x7hrs test, (-so based on my reading of the standard anyway it would appear that if there is adequate data available it is not necessary to do the temp test.)
Can someone with previous experience of this scenario please advise what the "inspection" and "available data" referred to above might involve. And which approach most people follow ..the inspection/data approach or the temp test? I'm guessing that it would involve design validation activities, e.g. mold flow simulations, inspection & dimensional analysis for sink, warpage etc. verification of tool construction to incorporate tool design guidelines to minimise residual stresses, etc, etc, but I'm just guessing and I'd really value the experience of someone who has gone down this road already preferably with a medical device incorporating plastic parts.
In terms of background the plastics involved are a range of PP, HDPE, LDPE, TPE and POM parts. The product is a medical device and the Mould Stress Relief test will form part of the design verification testing.
All comments, suggestions and insights welcomed, (my primary question is the main one I want answered the secondary question is a "nice to know".)
Kind regards,
SR