Hi,
I have a wireless medical device that measures a specific vital sign and transmits the data over to a dedicated software installed on a laptop.
The device is not a vital-signs monitor or life supporting device.
The device's essential performance is defined as the accuracy of the measured parameter.
During ESD gun testing at 15eV (a part of EMC 60601-1-2) the device may in some cases either:
(i) temporarily loses communication with the laptop, but automatically reestablishes the communication after few seconds and continues the measurement.
(ii) loses communication with the laptop, but requires to manually restart the measurement using the software
(iii) shuts down and requires to manually turn it on. Once turned on the measurement continues.
In all above-mentioned scenarios the software alerts the user that the data transmission is stopped and instruct the user if a specific action is needed (for example to turn on the device)
Although the accuracy of the measured vital sign (i.e. essential performance) is not affected during testing, the test lab claims that the device cannot pass the test, since there are cases in which the measurement does not restart automatically after the ESD interference.
In my opinion, their claim in not correct and is not supported by the standard. While it might be arguably considered as a functional fault, this is not an essential performance or safety fault and therefore cannot be a reason for a failing the test.
Before starting an argument with the test lab, is there anything that I am might missing here?
Your inputs will be much appreciated.
I have a wireless medical device that measures a specific vital sign and transmits the data over to a dedicated software installed on a laptop.
The device is not a vital-signs monitor or life supporting device.
The device's essential performance is defined as the accuracy of the measured parameter.
During ESD gun testing at 15eV (a part of EMC 60601-1-2) the device may in some cases either:
(i) temporarily loses communication with the laptop, but automatically reestablishes the communication after few seconds and continues the measurement.
(ii) loses communication with the laptop, but requires to manually restart the measurement using the software
(iii) shuts down and requires to manually turn it on. Once turned on the measurement continues.
In all above-mentioned scenarios the software alerts the user that the data transmission is stopped and instruct the user if a specific action is needed (for example to turn on the device)
Although the accuracy of the measured vital sign (i.e. essential performance) is not affected during testing, the test lab claims that the device cannot pass the test, since there are cases in which the measurement does not restart automatically after the ESD interference.
In my opinion, their claim in not correct and is not supported by the standard. While it might be arguably considered as a functional fault, this is not an essential performance or safety fault and therefore cannot be a reason for a failing the test.
Before starting an argument with the test lab, is there anything that I am might missing here?
Your inputs will be much appreciated.