for ISO/IEC 17025:2017 ; the following requirements should be considered :
7.8.2.1 Each report shall include at least the following information, unless the laboratory has valid reasons for not doing so, thereby minimizing any possibility of misunderstanding or misuse;
7.8.8.2 Amendments to a report after issue shall be made only in the form of a further document, or data transfer, which includes the statement “Amendment to Report, serial number... [or as otherwise identified]”, or an equivalent form of wording.
If the error can affect the results & the relevant issued reports ; it could be treated as change :
7.8.8.1 When an issued report needs to be changed, amended or re-issued, any change of information shall be clearly identified and, where appropriate, the reason for the change included in the report.
Such amendments shall meet all the requirements of this document.
7.8.8.3 When it is necessary to issue a complete new report, this shall be uniquely identified and shall contain a reference to the original that it replaces.