Roland chung
Trusted Information Resource
Hello everyone,
The clause 8.2.1 of 60601-1 3.1ed states that if a generic separate power supply is specified, then the specification in the accompanying documents is inspected.
Normally, the generic separate power supply doesn't provide with the medical device. It seems that the safety is guaranteed by the user manual only. But, the effectiveness of manual (as a risk control measures) is very weak, especially considering the reasonably foreseeable misuse. It is expected that 60950 grade, or worse power supplies (e.g. knockoff) may be used where 2 MOPP is required, especially for home-use devices (e.g. stimulators). So should 2 MOPP isolation be designed in the medical device when it is intended to be used with a generic sepatate power supply?
Could you please shed some light on this?
Thanks,
Roland
The clause 8.2.1 of 60601-1 3.1ed states that if a generic separate power supply is specified, then the specification in the accompanying documents is inspected.
Normally, the generic separate power supply doesn't provide with the medical device. It seems that the safety is guaranteed by the user manual only. But, the effectiveness of manual (as a risk control measures) is very weak, especially considering the reasonably foreseeable misuse. It is expected that 60950 grade, or worse power supplies (e.g. knockoff) may be used where 2 MOPP is required, especially for home-use devices (e.g. stimulators). So should 2 MOPP isolation be designed in the medical device when it is intended to be used with a generic sepatate power supply?
Could you please shed some light on this?
Thanks,
Roland