FelipeSchneider
Involved In Discussions
Hey everyone, it's widely acknowledged that the FDA has been requiring compliance with IEC TR 60601-4-2 for evaluating EMC performance criteria.
However, while going through IEC 60601-2-18 (which is relevant to the device I'm working on), I encountered this section:
Let's suppose my risk management concludes that an ESD event causing one disposable endoscope tip to burn is not an unacceptable risk because an operator can easily replace this "inexpensive" part. However, this would contradict the IEC TR.
So, my question is, which takes precedence? Can I disregard the IEC TR on this specific topic?
However, while going through IEC 60601-2-18 (which is relevant to the device I'm working on), I encountered this section:
202.6.1.10: The following shall not be considered unacceptable DEGRADATIONS for ENDOSCOPIC EQUIPMENT:
- the short interruption of illumination or image display, or resetting to 'standby' or 'safe' mode, when clearly indicated on the operation panel of the SUPPLY UNIT;
- if the RISK MANAGEMENT process shows that the DEGRADATION does not lead to an unacceptable RISK.
Let's suppose my risk management concludes that an ESD event causing one disposable endoscope tip to burn is not an unacceptable risk because an operator can easily replace this "inexpensive" part. However, this would contradict the IEC TR.
So, my question is, which takes precedence? Can I disregard the IEC TR on this specific topic?