I'm reviewing a test plan for monopolar ACTIVE ACCESSORY (with the ACTIVE ELECTRODE)... so not an ESU, not a NEUTRAL ELECTRODE, neither a power cable nor control mechanism. The active accessory will have no (electrical) energy within it except that provided by an ESU. This is new to me, so I have questions.
The particular IEC 60601-2-2:2017 A1:2023 is applicable, specifically it looks like this testing is appropriate:
XII. Electrical Safety and Electromagnetic Compatibility:
Additionally, endoscopic/laparoscopic electrosurgical instruments should demonstrate compliance with IEC 60601-2-18, Medical electrical equipment – Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipment. [...]
In addressing EMC, please be aware that according to IEC 60601-2-2, high frequency surgical instruments are intentional emitters of electromagnetic energy and, therefore, emissions testing to IEC 60601-1-2 is only needed for power generators in the idle state (i.e., powered on but surgical energy not activated). However, even during idle testing, particular attention should be paid to the effects of connected accessories and instruments, such as cord length (for resonant frequency) and instruments that contain electronics. For instruments with different cord lengths, connection types, or electronic components, it may not be appropriate to use a single “representative” instrument model for testing purposes.
If your submission is for a specific component of the electrosurgical device, you are still expected to evaluate your component while connected to other components of the electrosurgical device system. This should be performed consistent with how
Question: Aside from some consideration in the risk management file, is there any sort of OE (i.e. any meaningful test) for a monopolar accessory to demonstrate "consideration"? As a practical matter, there is a wide and deep pool of ESU that such a device could be connected to.
As an aside, the test plan includes testing for compliance to IEC 60601-1-2... Any suggestions what that would actually look like? I'm thinking that these sorts of accessories don't all of a sudden become troublesome antennae (in the 60601-1-2 sense) simply when connected... but this is new to me. If the ACTIVE ACCESSORY (not the cable) is covered in the ad hoc examples of Annex BD of 60601-2-2... I missed it. Is it that I am supposed to imagine a test where my non-energized ACTIVE ACCESSORY is (potentially) being energized by another, separate active acessory? If so, I feel like I can "risk amanagement file" my way out of such a circumstance.
The FDA guidance is equally clear and cloudy in XI. Performance Data E. System Testing. In particular:
1. Thermal Effects on Tissue
Per 60601-2-2 clause 201.11.1.2.1 APPLIED PARTS intended to supply heat to a PATIENT: ACTIVE ELECTRODES are considered to be APPLIED PARTS intended to supply heat to a PATIENT as part of their intended clinical effect (CUTTING and COAGULATION). Disclosure of temperatures and clinical effects is not required.... so this looks like the submission guidance asking for something that the standard explicitly doesn't want. I think a reasonable plan for submission (based on later text of the guidance doc) would be to do a like-for-like comparison on a variety of ex vivo animal tissues using representative modes of operation, yet I don't know how hard the regulatory reviewers would look at these analyses. Any suggestions? (I don't want to bring this up in a presub being completely uninformed!)
also 4. Capacitive Coupling
If the proposed device design includes monopolar electrosurgical electrodes for use in minimally invasive surgery, we recommend you test for active coupling resistance between the subject device and a conductive cannula/trocar device under simulated normal use conditions. This testing should be performed as recommended by the currently FDA-recognized versions of IEC 60601-2-18, Medical electrical equipment – Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipment and IEC 60601-2-2. For electrosurgical electrodes and accessories, we also recommend you address the risks of leakage currents through insulation as well as capacitive coupling with other surgical equipment (e.g., pulse oximeter probes, invasive blood pressure transducers, temperature probes camera systems) as a part of your risk management process (e.g., cross-coupling between different HF patient circuits).
Question: Is there any testing (supplimental or otherwise) in 60601-2-18 that is relevant for a device already being tested to 60601-2-2's 201.8.8.3.101 through 201.8.8.3.104? Obviously I don't want an ACTIVE ACCESSORY charging up a trocar, or anything else in the area (or patient!) that could be contacting a patient or user... I worry that I could be missing an important test in this area.
Another place 60601-2-18 is mentioned in the guidance:
XIV. Labeling
This appears pretty straightforward. but I have some confusion on the B. Device Labels section:
In your submission, we recommend that you provide illustrations to show how each component of your device is labeled to demonstrate that all controls, switches, and connections (including those for hand switched active electrodes and foot switches) are clearly, concisely, and permanently labeled to identify their function and other important information (e.g., Type BF Applied Part).
The nature of the device has me confused about whether or not the (monopolar) ACTIVE ACCESSORY (which fits the definition of an APPLIED PART) requires any of the following?
The particular IEC 60601-2-2:2017 A1:2023 is applicable, specifically it looks like this testing is appropriate:
- 201.8.8.3.101 Active Accessory insulation
- 201.8.8.3.102 Active Accessory HF leakage
- 201.8.8.3.103 Active Accessory HF dielectric strength
- 201.8.8.3.104 Active Accessory mains frequency dielectric strength
XII. Electrical Safety and Electromagnetic Compatibility:
Additionally, endoscopic/laparoscopic electrosurgical instruments should demonstrate compliance with IEC 60601-2-18, Medical electrical equipment – Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipment. [...]
In addressing EMC, please be aware that according to IEC 60601-2-2, high frequency surgical instruments are intentional emitters of electromagnetic energy and, therefore, emissions testing to IEC 60601-1-2 is only needed for power generators in the idle state (i.e., powered on but surgical energy not activated). However, even during idle testing, particular attention should be paid to the effects of connected accessories and instruments, such as cord length (for resonant frequency) and instruments that contain electronics. For instruments with different cord lengths, connection types, or electronic components, it may not be appropriate to use a single “representative” instrument model for testing purposes.
If your submission is for a specific component of the electrosurgical device, you are still expected to evaluate your component while connected to other components of the electrosurgical device system. This should be performed consistent with how
Question: Aside from some consideration in the risk management file, is there any sort of OE (i.e. any meaningful test) for a monopolar accessory to demonstrate "consideration"? As a practical matter, there is a wide and deep pool of ESU that such a device could be connected to.
As an aside, the test plan includes testing for compliance to IEC 60601-1-2... Any suggestions what that would actually look like? I'm thinking that these sorts of accessories don't all of a sudden become troublesome antennae (in the 60601-1-2 sense) simply when connected... but this is new to me. If the ACTIVE ACCESSORY (not the cable) is covered in the ad hoc examples of Annex BD of 60601-2-2... I missed it. Is it that I am supposed to imagine a test where my non-energized ACTIVE ACCESSORY is (potentially) being energized by another, separate active acessory? If so, I feel like I can "risk amanagement file" my way out of such a circumstance.
The FDA guidance is equally clear and cloudy in XI. Performance Data E. System Testing. In particular:
1. Thermal Effects on Tissue
Per 60601-2-2 clause 201.11.1.2.1 APPLIED PARTS intended to supply heat to a PATIENT: ACTIVE ELECTRODES are considered to be APPLIED PARTS intended to supply heat to a PATIENT as part of their intended clinical effect (CUTTING and COAGULATION). Disclosure of temperatures and clinical effects is not required.... so this looks like the submission guidance asking for something that the standard explicitly doesn't want. I think a reasonable plan for submission (based on later text of the guidance doc) would be to do a like-for-like comparison on a variety of ex vivo animal tissues using representative modes of operation, yet I don't know how hard the regulatory reviewers would look at these analyses. Any suggestions? (I don't want to bring this up in a presub being completely uninformed!)
also 4. Capacitive Coupling
If the proposed device design includes monopolar electrosurgical electrodes for use in minimally invasive surgery, we recommend you test for active coupling resistance between the subject device and a conductive cannula/trocar device under simulated normal use conditions. This testing should be performed as recommended by the currently FDA-recognized versions of IEC 60601-2-18, Medical electrical equipment – Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipment and IEC 60601-2-2. For electrosurgical electrodes and accessories, we also recommend you address the risks of leakage currents through insulation as well as capacitive coupling with other surgical equipment (e.g., pulse oximeter probes, invasive blood pressure transducers, temperature probes camera systems) as a part of your risk management process (e.g., cross-coupling between different HF patient circuits).
Question: Is there any testing (supplimental or otherwise) in 60601-2-18 that is relevant for a device already being tested to 60601-2-2's 201.8.8.3.101 through 201.8.8.3.104? Obviously I don't want an ACTIVE ACCESSORY charging up a trocar, or anything else in the area (or patient!) that could be contacting a patient or user... I worry that I could be missing an important test in this area.
Another place 60601-2-18 is mentioned in the guidance:
XIV. Labeling
This appears pretty straightforward. but I have some confusion on the B. Device Labels section:
In your submission, we recommend that you provide illustrations to show how each component of your device is labeled to demonstrate that all controls, switches, and connections (including those for hand switched active electrodes and foot switches) are clearly, concisely, and permanently labeled to identify their function and other important information (e.g., Type BF Applied Part).
The nature of the device has me confused about whether or not the (monopolar) ACTIVE ACCESSORY (which fits the definition of an APPLIED PART) requires any of the following?
- Testing beyond what is identified in 60601-2-2's 201.8.8.3.101 through 201.8.8.3.104 (I'm associating some of those tests with the conditions where the instrument NON-electrode portions are in contact with a patient and/or trocar, wether or not the accessory is energized)
- 60601-2-2 has this addition via clause 201.7.2.10 for APPLIED PARTS: The relevant symbols required for marking DEFIBRILLATION-PROOF APPLIED PARTS shall be attached to the front panel, but are not required on the APPLIED PARTS. Should the active accessory be labeled as a BF (or CF) part?... and if so what testing specifically demonstrates conformance?