renenatasha
Did you check on the Conformity Assessment Procedure (CAP) as well? IIb actual testing is needed if you choose EU type-examination to be your CAP.Think I found the difference in review of product documentation between IIa and IIb in Annex VII $4.5.3. Apparently for IIb a verification is needed, for IIa a confirmation is needed. I read this as for IIb the actual testing documentation will be reviewed, while for IIa the NB-review of a summary of testing is sufficient. What do you think?
- establish a test plan identifying all relevant and critical parameters which need to be tested by the notified body or under its responsibility in order to:
- — verify, for class IIb devices, the conformity of the device with the type described in the EU type-examination certificate and with the requirements of this Regulation which apply to those devices,
- — confirm, for class IIa devices, the conformity with the technical documentation referred to in Annexes II and III and with the requirements of this Regulation which apply to those devices;
Have you decided which approach to take, @ralf90 ?