I have a 400 mm scale that came from a SIP MU-214 measuring machine (vintage 1960). It is steel about 20x20mm square with a slightly recessed area on one side which has very fine engravings every mm.
The original calibration data says the max deviation for this scale was 0.7um, when calibrated by SIP in 1960.
I have sent this off to be calibrated by an ANAB accredited cal lab and they have given me a preliminary report of a reasonably gradual stretching of dimensions, ending at 400.019mm
Original SIP cal was at 20 C, the new cal was at 69F.
Is it possible a chunk of steel would grow by 19 microns in 400mm?
Seems like an awful lot.
The original calibration data says the max deviation for this scale was 0.7um, when calibrated by SIP in 1960.
I have sent this off to be calibrated by an ANAB accredited cal lab and they have given me a preliminary report of a reasonably gradual stretching of dimensions, ending at 400.019mm
Original SIP cal was at 20 C, the new cal was at 69F.
Is it possible a chunk of steel would grow by 19 microns in 400mm?
Seems like an awful lot.