Reports under change management | ISO 13485:2016 & ISO 9001:2015

#1
Hello All,

Do reports should be put under change management i.e. EC released into QMS?

Logically, reports (Software test reports, V&V test reports, etc.) should not be revised, since it is an output of any test done. Hence, review and sign off not on reports not enough? Do ISO 13485 and ISO 9001 specifically have the requirement?

Please share your thoughts.

Thanks in advance!
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
There are cases where records (like reports) can change and so change would need to be controlled.

Of course, all documents and records need to be controlled (protected, available at points needed, etc.) - which may be what your're indicating as released into QMS.
 

Tagin

Trusted Information Resource
#3
4.2.5 Control of Records states: "The organization shall document procedures to define the controls needed for the identification, storage, security and integrity, retrieval, retention time and disposition of records."

So its up to your organization to define the controls.
 

Tidge

Trusted Information Resource
#4
In my experiences, we've not placed reports (or other 1-and-done types of documents) under change management as they typically don't have a lifecycle like QMS documents do... a report typically won't 'revise', and it is unlikely to become 'obsolete'.

I can see using a change management system to collect approvals, but then you are stuck with having to figure out the states and approvals for transitioning the record to a different life-cycle state. Typically a document control system also fixes the hierarchy of a document at the time of approval as well (*1)

The primary reason why I would argue against "reports" going into a PDM system is that typically the records belong to a different class of "file", such as a Design History File (DHF) or a Device History Record (DHR).

(*1) Mature organizations will specify a meaningful hierarchy for reports/records, but it is "easy" for something like a DHF to be (re)organized at the whims of the file owner. Even DHRs can be shuffled around (not lost!) according to all sorts of whims (preference of the filers to have Serial Numbers in order, dates of PO filled, whatever)... and such 'ad hoc reorganizing' is generally a no-no once a document is put into change control. DHF and DHR contents fall more into the "retrieve the record" category. There is an additional concern that having all sorts of documents unattached to structures (QMS structures, or DMR/recipes) can have the appearance of a change control system 'out of control'.

EDIT: Catching myself and my use of DHR, DHF and DMR
 
Thread starter Similar threads Forum Replies Date
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
F Circuit board FAI reports- How do you folks deal with these? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
C Retention of QMS Plans and Reports ISO 13485:2016 - Medical Device Quality Management Systems 5
Y Examples of TRB Reports for MIL-PRF-31032 Qualification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
J Requested Validation plan and reports Manufacturing and Related Processes 4
B Vendor (Supplier) Evaluations and Reports Supplier Quality Assurance and other Supplier Issues 11
G Batch printing reports in Minitab Using Minitab Software 3
Watchcat Identification of Test Sample in Test Reports? Design and Development of Products and Processes 22
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Informational US FDA paper – Epidemiological Evidence on the Adverse Health Effects Reports in Relation to Mercury from Dental Amalgam: Systematic Literature Review Medical Device and FDA Regulations and Standards News 0
I Reducing CE marking cost using manufacturer test reports CE Marking (Conformité Européene) / CB Scheme 5
W Customer Signature / Acceptance Service Reports Service Industry Specific Topics 2
M CER (Clinical Evaluation Reports) updates - Product Codes EU Medical Device Regulations 2
Marc FDA safety scandal: 50K hidden reports of heart device malfunctioning - May 2019 Other US Medical Device Regulations 3
Q FDA 3500A Reports - Trip and Fall - Weight scale with a fold out platform Other US Medical Device Regulations 1
I AS9100 8.4.2 Type and Extent of Control - External provider test reports Manufacturing and Related Processes 24
M Informational USFDA – Radiological Health Regulations; Amendments to Records and Reports for Radiation Emitting Electronic Products; Amendments to Performance Stand Medical Device and FDA Regulations and Standards News 0
A Supplier Process Audits - Creation of Reports General Auditing Discussions 3
N How to resolve discrepancies in Level 3 PPAP supplier dimensional reports? APQP and PPAP 11
Marc vBulletin to Xenforo Import Problem Reports 28180928 Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 25
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
J AS9100 - Validation of Test Reports for Al 304 Raw Material AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R 510(k) - Finished Test Reports - RTA - Bit of a Complaint 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
P Medical Device Validation Protocols and Reports - Acceptance Criteria Other Medical Device and Orthopedic Related Topics 3
S What is the difference between Verification and Validation of Test Reports? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
I Where to submit Adverse Events Reports in Europe EU Medical Device Regulations 7
Q Writing Reports - Releasing Data into our Documentation System Document Control Systems, Procedures, Forms and Templates 1
J Internal Audits - Closing Audit Deficiency Reports (ISO 13485) Internal Auditing 4
A Process to Validate the Accuracy of Test Reports (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 38
R Why the FDA requests Lot Number in Testing Reports 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
B Existing reports and altering such reports AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Recognition of Certification Body Reports in different Countries Other Medical Device Regulations World-Wide 1
M Requirements, Verification Protocol and Reports All in One Document Design and Development of Products and Processes 6
R Medical Device Safety Assurance Case Reports Other Medical Device and Orthopedic Related Topics 6
C Using SAP for SCAR (Supplier Corrective Action Reports) Quality Assurance and Compliance Software Tools and Solutions 2
J Classifying Defect Reports from Business Partners US Food and Drug Administration (FDA) 4
P New to AS9100 Rev. C - Setting KPIs and FAI Reports AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
T Corrective Action Reports - Alternative Formats Nonconformance and Corrective Action 1
J Inspection Software to Balloon Drawings and Create First Article Reports Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
E Hardware (Such as Rivets and Bolts) - Chemical and Physical Test Reports AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
L CE Auditor Using ISO audit reports to show compliance to the MDD?? EU Medical Device Regulations 4
M FDA Acceptance of ISO 13485 Reports from a Recognized International Program Other US Medical Device Regulations 4
J Statistical Significance and SPC Control Chart Reports Statistical Analysis Tools, Techniques and SPC 9
D New to Gage Block Calibration Reports Measurement Uncertainty (MU) 6
P Internal Audit Reports reviewed by management? 21 CFR 820.22 Interpretation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
C The prudence of combining non-conformance reports with corrective/preventive action Nonconformance and Corrective Action 10
G Device Output (Reports) - Any regulatory requirements? US Food and Drug Administration (FDA) 4
K Process Owners signing Internal Audit Reports Internal Auditing 6
K Evidence Documentation in Internal Audit Reports Internal Auditing 24
L Clarification on Clinical Evaluations and Clinical Evaluation Reports Other Medical Device and Orthopedic Related Topics 5

Similar threads

Top Bottom