Our device uses a transformer to isolate the patient connection from the rest of the circuits. It provides 2 MOPP from primary to secondary and the frame is earthed.
This device provides high voltage outputs so the primary circuits are connected to a (intentionally) high impedance voltage source and the output transformer then steps up the very brief pulses.
If one uses the concept of the working voltage being the sum of the peak voltages in the two windings then the answer is very high with the consequence that the requirements for dielectric strength test voltage are huge (in the order of 8kV).
Is there a better way of working out the working voltage?
I'm also struggling to work out the requirements for the turn-layer tests (15.5.2) for a pulse transformer where the duty cycle is so low. If any one has ideas on that I would be grateful.
Further question:
The pulse transformer provides 2 MOPP from the primary to the (applied part) secondary. But what is the requirement for between the secondary and the earthed frame? Does the earth of the frame count as 1 MOPP? so that the insulation just needs to provide 1 MOPP? (consider the insulation examples given in Annex J figure J.6 and J.7).
Many thanks for any suggestions.
This device provides high voltage outputs so the primary circuits are connected to a (intentionally) high impedance voltage source and the output transformer then steps up the very brief pulses.
If one uses the concept of the working voltage being the sum of the peak voltages in the two windings then the answer is very high with the consequence that the requirements for dielectric strength test voltage are huge (in the order of 8kV).
Is there a better way of working out the working voltage?
I'm also struggling to work out the requirements for the turn-layer tests (15.5.2) for a pulse transformer where the duty cycle is so low. If any one has ideas on that I would be grateful.
Further question:
The pulse transformer provides 2 MOPP from the primary to the (applied part) secondary. But what is the requirement for between the secondary and the earthed frame? Does the earth of the frame count as 1 MOPP? so that the insulation just needs to provide 1 MOPP? (consider the insulation examples given in Annex J figure J.6 and J.7).
Many thanks for any suggestions.