When does an engineering change become a design change?

J

Jon Griver

#1
Dear Forum,

Manufacturing companies generally control engineering changes using ECO forms or similar mechanisms.

It is not clear to me how it should be decided whether an engineering change should be managed through the 'full' Design Control process, i.e. including design planning, design input, etc. etc. Should the design control documentation be reviewed during every ECO?

What about the situation where a product was developed before Design Control was implemented, and all changes since then have been managed by ECO's, without design control documentation ever being prepared.

And finally, does anyone know what would be the FDA's approach to this issue?

Jon Griver
 
Elsmar Forum Sponsor

Al Rosen

Staff member
Super Moderator
#2
Jon Griver said:
Manufacturing companies generally control engineering changes using ECO forms or similar mechanisms.

It is not clear to me how it should be decided whether an engineering change should be managed through the 'full' Design Control process, i.e. including design planning, design input, etc. etc. Should the design control documentation be reviewed during every ECO?
I believe that all changes should be reviewed to determuine the extent of the controls required. The FDA regulations say that you have to perform verification and validation "where appropriate". If you do not perform verification or validation, they require documented justification for not doing it. As part of an inspection they will review engineering changes, especially software changes. I had the pleasure of an inspection last year.

Jon Griver said:
What about the situation where a product was developed before Design Control was implemented, and all changes since then have been managed by ECO's, without design control documentation ever being prepared.

And finally, does anyone know what would be the FDA's approach to this issue?
Even though the device was developed prior to the design control regulations, when the regulations went into effect, you were required to institute the controls. Simply stated, they will issue your company a warning letter. You can search some of the warning letters issued, at this web site:
http://www.fda.gov/foi/warning.htm

To avoid the warning letter, you will need to implement corrective action prior to an inspection. If not, then at least before the end of the inspection. Letting it go until the inspection, the finding will be documented on the 483 and might still make it to a warning letter depending on what else they uncover.
 

Wes Bucey

Quite Involved in Discussions
#3
Jon Griver said:
Dear Forum,

Manufacturing companies generally control engineering changes using ECO forms or similar mechanisms.

It is not clear to me how it should be decided whether an engineering change should be managed through the 'full' Design Control process, i.e. including design planning, design input, etc. etc. Should the design control documentation be reviewed during every ECO?

What about the situation where a product was developed before Design Control was implemented, and all changes since then have been managed by ECO's, without design control documentation ever being prepared.

And finally, does anyone know what would be the FDA's approach to this issue?

Jon Griver
All our Engineering Change Orders (to pass muster with the FAA) go through a process to determine the "impact" of the change. I redacted a computer form (to remove company ID) and printed it to show the checklist we consider for the process. Some changes have little or no impact and are dealt with accordingly. Others may have great impact and go through a process similar to a new design.

A non-impact change might be "changing fasteners in a non-user-accessible portion of a device from phillips head to star drive to accommodate assembly machines." It is important to note there is a point where the change is so drastic, it is no longer viable to consider it a change, but a new design, effectively obsoleting the original design.

I hope this helps you put your own changes into proper perspective and to help you form a "justification" when you do not go through a new verification and validation of the revised design.

Any questions about the form can be addressed in this thread.
 

Attachments

Al Rosen

Staff member
Super Moderator
#4
Other aspects I did not mention in my previous post involve review to determine what effect the change has to safety and whether a risk analysis should be updated as well as a determination of the need for a Premarket notification, 510(k) to be submited.

See this post for the FDA design control guidance document
http://elsmar.com/Forums/showpost.php?p=72191&postcount=44

and, see the guidance document I've attached regarding 510(k)s for a change to a device.
 

Attachments

J

Jon Griver

#5
Thank you Al and Wes for your replies.

I would like to summarize what I have learned from your replies and other research I have done.

I previously regarded the design change mechanism and the ECO mechanism as parallel routes to implement changes, where design change is for 'big' changes and ECO is for 'small' changes.

My viewpoint now is that ALL changes have to be evaluated with regard to the design change mechanism, though not all elements of design change will be applicable to all changes. The ECO mechanism is actually a document control procedure, basically describing how to fill out the form, and does not really relate to the technical content of the change.

I would welcome any comments on this.

Regards,


Jon
 

Wes Bucey

Quite Involved in Discussions
#6
Jon Griver said:
Thank you Al and Wes for your replies.

I would like to summarize what I have learned from your replies and other research I have done.

I previously regarded the design change mechanism and the ECO mechanism as parallel routes to implement changes, where design change is for 'big' changes and ECO is for 'small' changes.

My viewpoint now is that ALL changes have to be evaluated with regard to the design change mechanism, though not all elements of design change will be applicable to all changes. The ECO mechanism is actually a document control procedure, basically describing how to fill out the form, and does not really relate to the technical content of the change.

I would welcome any comments on this.

Regards,


Jon
By Jove! I think he's got it!

Excellent synopsis, Jon! Especially the point about document control. Google "configuration management" for more detailed discussions of a subset of document control which may discuss the interchangeability of various versions of documents and products. (as in my example of merely changing drive heads on fasteners while other aspects - thread, length, material - remain the same)
 
A

Al Dyer

#7
I know in the auto arena that design responsibility comes into play when an entiry has the authority to change the design record.

Is there a similar activity in place in the FDA/Medical field?

Al...
 

Al Rosen

Staff member
Super Moderator
#8
Al Dyer said:
I know in the auto arena that design responsibility comes into play when an entiry has the authority to change the design record.

Is there a similar activity in place in the FDA/Medical field?

Al...
Al:

In medical devices that would be the entity that holds the FDA approval for the device.
 
Thread starter Similar threads Forum Replies Date
J Definition What does Engineering Judgement mean in MSA? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
Chennaiite Value Engineering - Does Value Engineering really add value? Preventive Action and Continuous Improvement 3
C Does all gauges mean every gage? Manufacturing Engineering (Paragraph 7.6.2) IATF 16949 - Automotive Quality Systems Standard 10
S Does Quality Engineering have a role in QFD (Quality Function Deployment)? APQP and PPAP 1
J What exactly does 4.2.3.1 "Engineering Specifications" in ISO/TS 16949 mean? IATF 16949 - Automotive Quality Systems Standard 22
R 4.2.3.1 Engineering Specifications - Intent? What does this clause apply to? IATF 16949 - Automotive Quality Systems Standard 7
lanley liao Does all of the suppliers need to integrated into the supplier list qualified of the company? Oil and Gas Industry Standards and Regulations 2
0 To which part of 13485 does this refer? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
B Does anybody know how to get older versions of Minitab to work in Windows 10? Quality Tools, Improvement and Analysis 9
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 3
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Brexit When does the UK responsible person need to be in place? UK Medical Device Regulations 10
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
Watchcat Does "Similar Device" = "Predicate"? EU Medical Device Regulations 7

Similar threads

Top Bottom