Acceptable calibration accuracy of a 60" linear measuring device

qcman

Registered Visitor
#1
With typical length tolerances of ± .015 / 0.39mm what would be an acceptable calibration accuracy of a 60" linear measuring device?
 
Elsmar Forum Sponsor

qcman

Registered Visitor
#3
That is what I thought as I do not see a need for 10x with the intended use. Do you spread that 4x over the entire 60"? In other words I am checking every 5" so would the 4x be every 5"? I have always outsourced the calibration of these units believing they were actually being calibrated :mad: Anyway I have purchased grade blocks and will bring the calibration in house hence the questions.
 

dwperron

Trusted Information Resource
#4
I would need more information on the tolerance before I could give an answer.

First, if it is a "typical" tolerance of ±0.015" that is what it should be, not what it must be. You can't really call a "typical" device out of specification since it is not a warranted tolerance...

Also, assuming that this might be a warranted tolerance, is it over the full range of the device (60")? Or is there also a linearity specification (such as 0.005" per 12") in addition.
 

Al Rosen

Holed-up in a Hotel in South Florida
Staff member
Super Moderator
#5
That is what I thought as I do not see a need for 10x with the intended use. Do you spread that 4x over the entire 60"? In other words I am checking every 5" so would the 4x be every 5"? I have always outsourced the calibration of these units believing they were actually being calibrated :mad: Anyway I have purchased grade blocks and will bring the calibration in house hence the questions.
I would Calibrate it where you are using it on the scale. Measuring +/-.015" over 60" with a steel rule is tight.
 

qcman

Registered Visitor
#6
Maybe I need to better clarify. We cut a variety parts to different lengths with most having a tolerance of ± .015. I checked the linear every 5" and it measures within .00047 or below up to 50". 55" measures at +.001 which is more than acceptable for our process.
 

Ninja

Looking for Reality
Trusted Information Resource
#7
55" measures at +.001 which is more than acceptable for our process.
Pretty sure you just answered your own question...with the "more than acceptable for our process" statement....however...

Math says +0.001" = 0.0254mm (assumed inches since you didn't state units on 0.001)

Tolerance 0.015mm / 4 = 0.004mm

Math says it's not good enough...assuming I assumed units correctly.
 

Al Rosen

Holed-up in a Hotel in South Florida
Staff member
Super Moderator
#8
Pretty sure you just answered your own question...with the "more than acceptable for our process" statement....however...

Math says +0.001" = 0.0254mm (assumed inches since you didn't state units on 0.001)

Tolerance 0.015mm / 4 = 0.004mm

Math says it's not good enough...assuming I assumed units correctly.
His stated tolerance in the op is +/-.015 inches, not mm. I think qcman is stating it in inches.
 

Ninja

Looking for Reality
Trusted Information Resource
#9
Yup, looks so...

With typical length tolerances of ± .015 / 0.39mm
Didn't do the math on that one to realize it was inches/mm. Thanks.

Thus Math says +0.001"

Tolerance 0.015" / 4 = 0.004"
If it is indeed +/- 0.015"...the math looks better.

On the edge of 4x...but the statement that it more than meets needs rules (until a customer disagrees).
 

qcman

Registered Visitor
#10
Yes inches. I am trying to drop english all together here but it is incredible how much resistance there is to this :confused: Anyway that would be .015 / 4 = .0037 inches. This may sound like a dumb question but would that be .0037 for any given zone checked but not to exceed .0037 over 60"? Btw this is a Mitutoyo KA-200 counter with a AT715 scale which should have an accuracy of 7 micrometers / 0.00027 inches.
 
Thread starter Similar threads Forum Replies Date
A Is "0" an Acceptable Calibration Point? General Measurement Device and Calibration Topics 13
C Is a Post-it acceptable as a Calibration Label? General Measurement Device and Calibration Topics 15
G Primary vs. Secondary Standards - Are secondary standards acceptable for calibration? General Measurement Device and Calibration Topics 9
F Calibration Issue - Continuous Temperature Monitoring System was not Acceptable General Measurement Device and Calibration Topics 7
C Acceptable Testing Range for Calibration Studies on a Measurement Device General Measurement Device and Calibration Topics 3
S How to Set Acceptable Calibration Limits for Measurement Instruments? General Measurement Device and Calibration Topics 11
S Calibration Labels Per Piece or is a Listing Acceptable? General Measurement Device and Calibration Topics 4
J What is an acceptable % of overdue gages not turned in for calibration? General Measurement Device and Calibration Topics 7
C SPC for calibration check data? Daily gram scales check - What is acceptable? Statistical Analysis Tools, Techniques and SPC 5
S Is a low calibration TUR acceptable in the eyes of ISO17025 and TS16949? 4:1 TUR Rule ISO 17025 related Discussions 2
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
M Are Fungal counts acceptable in class 1000 clean rooms? Other ISO and International Standards and European Regulations 2
T ISO 14971-2019 doubt - Evaluate if estimated risks are acceptable ISO 14971 - Medical Device Risk Management 9
samer Acceptable limits for Spills - Tracking Hydraulic Spills ISO 14001:2015 Specific Discussions 3
V Acceptable maximum RSD (relative standard deviation) for an sample size Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
F It is acceptable moving remote locations staff to manufacturing plant for auditing? IATF 16949 - Automotive Quality Systems Standard 3
Q Is it acceptable to mix components from two different lots into an assembly? ISO 13485:2016 - Medical Device Quality Management Systems 1
C Acceptable NDC for %GR&R part inspection to Tolerance (%Tolerance) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Risk Register - have we considered enough and is the format acceptable? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
K Operator Checks - How to show that they were completed on a checklist as acceptable IATF 16949 - Automotive Quality Systems Standard 2
C Example Work Instruction/Procedure for AQL (Acceptable Quality Level) AQL - Acceptable Quality Level 4
S Dates on Labels acceptable to the USA - GS1 General Specification 3.4.4 Other US Medical Device Regulations 3
E Root Cause Analysis - Is Insufficient Understanding an acceptable Root Cause? General Auditing Discussions 9
W Unattached (stand-alone) Forms acceptable for AS9100C? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
V Why Gauge R&R 10% is acceptable for variable instruments ? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
X Acceptable methods to store and archive records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Is it acceptable to audit to the Nuclear Principles? Quality Manager and Management Related Issues 9
I Is it acceptable it is to list new products under an existing 510(k)? Other US Medical Device Regulations 14
S If this EN ISO 11137 certificate acceptable for Contract Sterilization? ISO 13485:2016 - Medical Device Quality Management Systems 2
D What is satisfactory & acceptable for Coliform - Staphylococcus - Hands and Surfaces Food Safety - ISO 22000, HACCP (21 CFR 120) 3
S Is a Dynamic Controlled Form acceptable per 21CFR820 Document Control Systems, Procedures, Forms and Templates 3
C Lowest Acceptable Grade for Master Gage Set Calibration and Metrology Software and Hardware 1
B Digital Signature Solutions acceptable to use in an FDA Medical Device Environment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Is a ndc less than five is also acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T Technical regulations that determine the acceptable limits of cigarettes General Measurement Device and Calibration Topics 4
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is it acceptable to sterilize products without claiming and labeling "sterile" ? ISO 13485:2016 - Medical Device Quality Management Systems 6
M Quality Objectives - Acceptable Levels and Functions ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 37
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Is embedding documents within a Contract an acceptable practice ? Contract Review Process 9
B Non Acceptable % Tolerance Result - Decrease Sorting Limits Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
F ANSI ASQ Z1.4 - Is the following process acceptable? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C Maximum acceptable Tolerance - Uncertainty Ratio (TUR) General Measurement Device and Calibration Topics 2
S Is it acceptable to have an ASL for Critical suppliers alone? Supplier Quality Assurance and other Supplier Issues 4
S Is this Quality Policy acceptable in the context of ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 24
A Help with acceptable screw gage values per standard BS 919 for Go/ No Go gage General Measurement Device and Calibration Topics 4
S What are typical Acceptable Quality Levels (AQL) adopted in Food industry? Food Safety - ISO 22000, HACCP (21 CFR 120) 1

Similar threads

Top Bottom