Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance

greif

Involved In Discussions
#1
We have a x-y coordinate machine that has had an uncertainty assigned to it, based on length of measurement k=2 0.0042mm + (0.0000135L) where L is in mm.
example; at 0 mm the uncertainty is 4.2um, at 400 mm the uncertainty is 9.6um.

We periodically calibrate/verify this machine by reading a master calibrated scale.

Question: is it proper to use the full above mentioned uncertainty as my verification tolerance? Or do I need to include some other factor?

Thanks!!
 
Elsmar Forum Sponsor

dwperron

Trusted Information Resource
#2
You need to check with the manufacturer for the correct interpretation of the specification.
These days too many companies use accuracy, tolerance, and uncertainty interchangeably.

If they really mean uncertainty they will be able to state what uncertainty contributors are included, such as resolution, repeatability, hysteresis, environmental, etc. along with the basic accuracy specification.
If it is basic instrument accuracy, then you need to include the other contributors to determine the uncertainty of the machine.

For the purpose of calibrate/verify the machine, you would use the basic accuracy specification.
 

Ninja

Looking for Reality
Trusted Information Resource
#3
Additionally, consider the accuracy/reliability/uncertainty (not interchangeable, but not additive either)...that you NEED.

One question is "Is the gage good as new?"...a whole different question is "Is this good enough for my needs?"
Answer both questions...
 

greif

Involved In Discussions
#4
My main focus is to accurately report uncertainty in reports that go to customers (getting set up to be 17025 accredited). The machine is quite old and does not have a useful tolerance from the former manufacturer.
So my original question; can I use the uncertainty that we calculated in -house as my machine periodic verification tolerance, or is that too circular?
 

dwperron

Trusted Information Resource
#5
For the full uncertainty calculation that you report you will need include your "calibrate/verify" results (the accuracy that you determined the machine is capable of), the uncertainty of your calibration standard(s), any environmental factors, the resolution of your results, the repeatability / reproducibility of your results (based on your studies), and any other significant contributors in your process.

As long as your standard(s) all have accredited calibrations you are not too circular.
 
Thread starter Similar threads Forum Replies Date
T Formal Q Self Assessment - Problem with assigning Product and Process Customer and Company Specific Requirements 1
K Assigning the Correct Procedures Applicable to Job Titles General Auditing Discussions 15
K IEC 62304 Amd 1 2015 - Figure 3 – Assigning Software Safety Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
Q Corrective Action - Assigning KPIs to Non-Guilty Processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V What is the criteria for assigning the new-revision number for an FMEA? FMEA and Control Plans 6
Q Assigning Responsibilities - Responsible Department or positions? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
P Assigning Occurrence level in FMEA ISO 14971 - Medical Device Risk Management 7
AnaMariaVR2 Method or System for Assigning Project File Names US Food and Drug Administration (FDA) 10
J Determining the Appropriate Sampling Regime when Assigning Values to an IVD Device Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
M Assigning a PDF for Analog Resolutions - OIML R49-2 Test Methods for Cold Water Meter Measurement Uncertainty (MU) 3
B Assigning Targets for Programs in a New Company Miscellaneous Environmental Standards and EMS Related Discussions 16
ScottK Assigning yourself more power and authority. Career and Occupation Discussions 4
S Assigning Internal Part Number Document Control Systems, Procedures, Forms and Templates 4
D Assigning numbers to control plans etc. FMEA and Control Plans 7
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0

Similar threads

Top Bottom