Hi all,

I have a fluke 753 dpc and i decide to send in to fluke for calibration.

But from the sample cert they send me, fluke only calibrate thermocouple type k measure and source.

My question here is if i use type T for measure and source, and when i issue accredited cert to my customer, is my cert still traceable to fluke cert? Is it still consider iso 17025?

Thanks.
 
Elsmar Forum Sponsor

Ninja

Looking for Reality
Staff member
Super Moderator
#2
Re: traceability question

If Fluke will not use a Type T for calibration if you request it, perhaps you would be better off sending it somewhere else...

Something to consider...
 

BradM

Staff member
Admin
#3
Basically what they are doing is verifying the accuracy of the meter for thermocouples using one curve. There are multiple different curves in the meter (K, J, T, N, S, etc.) So if the millivolt and cold junction compensator is verified to be accurate, then it's just an electronic curve thing to switch from T/C to T/C.

What you should do is contact Fluke and ask if they can check your meter with Type T instead of K.

I am intrigued regarding your issuing an accreditation certificate to your customer. Are you using the Fluke meter to calibrate something for your customer? The Fluke 753 is a process meter; and I wouldn't consider it suitable for calibrating much; especially temperature.
 
N

neoth

#4
Hi Brad,

Yes i am using this fluke 753 to calibrate my customer type T temperature indicator using simulation method.

I tried I to contact fluke to use type T but they told me it is a standard procedure and they cant change it.

So can my issued cert be traceable to fluke cert if i am using T?
 
G

Gkokal

#5
Hi Neoth,

The thermocouple functions on process calibrators work on a voltage generated from the two different kinds of metals used in the thermocouple production. If the type k thermocouple measure function passes calibration on the cert, then type T will work just fine. Same for source. It is a voltage signal being tested, not actual temperature.

The thermocouple itself is a different standard, so watch yourself if you are measuring actual temperature.
 
Thread starter Similar threads Forum Replies Date
A Fluke Digital MultiMeter Calibration Traceability and Certificate General Measurement Device and Calibration Topics 11
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
T In house Calibration? Measurement equipment traceability requirements - ISO9001 General Measurement Device and Calibration Topics 12
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
A Is NIST traceability required if a Refractometer calibration is done in-house? Calibration Frequency (Interval) 6
P Importance of Calibration Traceability to the SI General Measurement Device and Calibration Topics 1
R Torque Tester Calibration Traceability and Certificate question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Pipette Calibration & ISO 8655 - No NIST Standard(s) Traceability ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
W Calibration Register - Traceability - Data Transfer to a New System General Measurement Device and Calibration Topics 7
D Calibration Traceability meet AS9100C Requirement? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
T Class "F" Weight "Certificate of Traceability" Requirements - Calibration General Measurement Device and Calibration Topics 7
D Equipment calibration standard traceability - The basis used for calibration General Measurement Device and Calibration Topics 5
R Outside Calibration - Accredited, certified - Is NIST traceability guaranteed? General Measurement Device and Calibration Topics 8
A Calibration traceability - Profile projector General Measurement Device and Calibration Topics 3
Marc Calibration is the process of establishing a KNOWN RELATIONSHIP - Traceability General Measurement Device and Calibration Topics 0
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 2
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 1
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 4
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3

Similar threads

Top Bottom