We are facing a major ESD test challenge with integrating a high-quality image sensor into a class-2 product. We spent 5 months changing our optical and mechanical designs without getting to the required test levels thus far. We can not mechanically block the optical path due to the loss of video quality, and we can only get about +/-6KV with thin transparent windows. If the product remains safe, and it does not cause any harm to the patient or operators, can the manufacture of a class-2 medical device define a lower ESD test level for a sensitive imaging sensor that can not pass +/-8KV or +/-15KV static discharge? Is this managed in the risk management document? Should there be a specific instruction for the certified test labs to exclude a specific area during the 60601 ESD test? If this is permitted, does this need special labeling?