For the sake of discussion, the OOT of tolerance condition is in the neighborhood of 10 microns while our tightest part tolerance is no less than +/- 0.5 mm. Would it be acceptable to simply state the ratio between the OOT condition and the minimum customer tolerance meets a 50:1 ratio and therefore would not influence the stated results of measurements taken prior to the roamer arm calibration?
My opinion.... yes. Write this up to demonstrate that the amount of error assessed at calibration time was insignificant in terms of potential process impact ( 10 microns versus .5mm) and thus no potential process risk. Or... however the language is acceptable within your industry/ organization.
Like Paul mentioned, I would invest in seeing if you can avoid future calibration failures. Maybe see if you can establish a suitable acceptance tolerance for the roamer arm. So mfg. tolerance is... ±10 micron but your process acceptance tolerance is say... ±.05mm. Make sure the calibration vendor adjusts the arm to maximum accuracy (remove error) even if it is found within specifications. Otherwise that error can build and become a failure in the future.