I was wondering if anyone can help regarding how to specify ECG leads during electrical safety testing.

If I have an ECG device with 3 leads (RA, LA and LL), should it matter how these are specified on a 60601 safety tester? For example, provided that I am defining all leads as CF, should it matter which inputs (RA, RL, LA, LL etc...) I am plugging them into on the tester?

I am finding different patient auxiliary current measurements depending upon how they are being specified, but cannot work out why this should be the case.

Many thanks,
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
There might be a small amount of patient auxiliary current coming from the "right leg drive", this is kind of like noise cancellation.

For 3 lead systems, this can be set up to be the non-used lead. For example, if the patient monitor is displaying "Lead I", this is actually the voltage between Lead electrode LA and RA, and LL is then used for noise cancellation. So you might measure a small amount of current from LL.

If you switch to Lead II on the display, this is LL-RA which means LA has the noise cancelling function.

Another possibility is the use of higher frequency currents for detecting "leads off" condition. If your safety tester has the function to turn off the 1kHz filter, turn it off and see if there are higher currents indicated.

Anyway, I would discuss with the designer or manufacturer to find the source of the auxiliary currents if you want to determine the worst case.

The mains sourced leakage (capacitive leakage from mains parts) should be the same for all leads (and also all leads together). They are all on the same circuit so there's no difference.

For the mains on applied part test, in theory it is the same as for other mains related currents. But the position of the cable can change the result. So you might be getting different results due to moving the cable around around without realising.
 

Roland chung

Trusted Information Resource
#3
For the mains on applied part test, in theory it is the same as for other mains related currents. But the position of the cable can change the result. So you might be getting different results due to moving the cable around around without realising.
Could you please explain more about the effect of the cable? Thank you!
 
#4
There might be a small amount of patient auxiliary current coming from the "right leg drive", this is kind of like noise cancellation.

For 3 lead systems, this can be set up to be the non-used lead. For example, if the patient monitor is displaying "Lead I", this is actually the voltage between Lead electrode LA and RA, and LL is then used for noise cancellation. So you might measure a small amount of current from LL.

If you switch to Lead II on the display, this is LL-RA which means LA has the noise cancelling function.

Another possibility is the use of higher frequency currents for detecting "leads off" condition. If your safety tester has the function to turn off the 1kHz filter, turn it off and see if there are higher currents indicated.

Anyway, I would discuss with the designer or manufacturer to find the source of the auxiliary currents if you want to determine the worst case.

The mains sourced leakage (capacitive leakage from mains parts) should be the same for all leads (and also all leads together). They are all on the same circuit so there's no difference.

For the mains on applied part test, in theory it is the same as for other mains related currents. But the position of the cable can change the result. So you might be getting different results due to moving the cable around around without realising.

Hi Peter,

I have a question. For a battery powered device, should we measure the the auxiliary between the battery terminal and a real patient connection port in single fault condition? For example, a device powered by 2 AA batteries and also have a real patient connection port. In single fault condition, a return loop may form between the patient and one AA battery. As the patient resistance is 1K ohm, the patient auxiliary current will be 1.5V/1000Ohm =1.5mA (This exceeds the limit in 60601-1 Ed. 3.2)

Thanks and regards,
 

Peter Selvey

Staff member
Super Moderator
#5
The path should be considered but in practice not tested because the battery terminals should be reasonably insulated from the patient's applied parts/patient circuit. It's not simply a matter of measuring from the terminals because they are there, there has to be a plausible scenario involved to get the voltage to the patient. It could happen, but it's not difficult to avoid by sensible design.

I know that 601-1 has some crazy limits for cr/cl like 3.4mm, 500V which could see some designs fail even though they are perfectly safe. The 500V/3.4mm are secondary circuits that are derived from mains circuits, the standard is being lazy for MOPP and just wants one limit for all. In that case it might require some bending of the standard to get common sense e.g. 1mm/50V is fine for a battery operated device, and less if the distance is fixed e.g. on a PCB.
 
#6
The path should be considered but in practice not tested because the battery terminals should be reasonably insulated from the patient's applied parts/patient circuit. It's not simply a matter of measuring from the terminals because they are there, there has to be a plausible scenario involved to get the voltage to the patient. It could happen, but it's not difficult to avoid by sensible design.

I know that 601-1 has some crazy limits for cr/cl like 3.4mm, 500V which could see some designs fail even though they are perfectly safe. The 500V/3.4mm are secondary circuits that are derived from mains circuits, the standard is being lazy for MOPP and just wants one limit for all. In that case it might require some bending of the standard to get common sense e.g. 1mm/50V is fine for a battery operated device, and less if the distance is fixed e.g. on a PCB.
Thank you Peter for your prompt reply. This is really helpful.
 
Thread starter Similar threads Forum Replies Date
H ECG Modified Lead Placement - IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Preparing the Technical File for a 12 Lead ECG Monitor EU Medical Device Regulations 8
A 60601-1 : Integrated Dry ECG Electrodes = 2 Patient connections inside 1 applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Insulation diagram for ECG monitoring and diagnosis system. IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
R ECG Risk Analysis Standards ISO 14971 - Medical Device Risk Management 2
W Isolation requirements for Wireless ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E ECG device without ECG REPORTS - IEC 60601-2-25: 2011 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S IEC 60601-2-25 ECG 4 electrodes and derived leads IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1-2 Radiated Immunity Test - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1-2 immunity to proximity fields Test setup - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1-2 - RF Conducted Immunity Test Setup - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 Radiated E-Field Emission Testing - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
kevinkre IEC EN 60601-2-25 subclause 201.12.4.106.2 Channel Crosstalk (ECG Multichannel Crosstalk) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Indication of Inoperable ECG test result IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K Is Defib Protection mandatory for wearable ECG? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
S Testing my ECG device as per IEC 60601-2-25 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Standards Specific to Single Channel ECG Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Old software and EN 62304 - ECG software for the display of ECG IEC 62304 - Medical Device Software Life Cycle Processes 2
R Classification of ECG Module IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
R Requirements for an ECG device with GSM for Data Transmission IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Dielectric Test for Ambulatory ECG Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A China's Standards for ECG Machines China Medical Device Regulations 0
A How to purchase CTS-ECG Database Quality Assurance and Compliance Software Tools and Solutions 1
P IEC 60601-2-25:2011 Comparison Table (ECG, Diagnostic) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S AAMI EC57 for ECG Medical Device - Seeking 3rd Party Test Laboratory US Food and Drug Administration (FDA) 5
A EN 60601-2-51 applicability to an ECG Event Recorder IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
A Validity of IEC 60601-1 and IEC 60601-2-27 - An intermittent ECG event recorder IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Classification of ECG/EKG Electrodes according to MDD 93/42/EEC? EU Medical Device Regulations 2
D ECG Electrodes - Duration of Use (How long they are expected to stay on the patient) Other US Medical Device Regulations 2
D Independent Testing of ECG Electrodes in accordance with AAMI standards Reliability Analysis - Predictions, Testing and Standards 5
G Medical Device Accessories (e.g.: ECG Patient Cables) - Traceability Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
bio_subbu FDA Issues Guidance Document on Electrocardiograph (ECG) Electrodes Other US Medical Device Regulations 0
J IEC EN 60601-2-27 subclause 50.102.5 ECG Multichannel Crosstalk IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
S Does an ECG extension cable need a 510k? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R EC11 - Guidance Document - FDA 510 (k) Submission - ECG Monitors 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
R Not every INCIDENT report will lead to a corrective action EU Medical Device Regulations 6
B Lowest cost way to pass Lead Auditor exam ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
R Through-Hole Electronics - Checking Lead Length Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
M Tips on preparing for IATF 16949 Internal Lead Auditor exam Manufacturing and Related Processes 1
Gun46 ISO 9001 : 2015 Lead Auditor Exam General Auditing Discussions 16
S ISO 13485 Lead Auditor - Debate between our Quality Team and Regulatory Auditor - Internal Auditor Training ISO 13485:2016 - Medical Device Quality Management Systems 40
R ISO 45001 Lead Auditor Exam paper Training - Internal, External, Online and Distance Learning 1
T Qualification testing of Lead acid batteries Reliability Analysis - Predictions, Testing and Standards 0
qualprod Monitoring of lead time - Good KPI identification? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
W Certification for IATF Lead Auditor will expire in 2020 IATF 16949 - Automotive Quality Systems Standard 2
S IRCA Lead Auditor training and Exam tips Training - Internal, External, Online and Distance Learning 5
J Getting training either from ASQ or from SAI Global - ISO 9001 Lead Auditor training Training - Internal, External, Online and Distance Learning 1
F Position of Quality Manager - Can the organization's Quality Lead be shown as Quality Manager of the lab as well? ISO 17025 related Discussions 5

Similar threads

Top Bottom