bio_subbu
Super Moderator
Hi PriyaSir
We have claimed that 7.5.1.2.1 is not applicable
a) product is NOT cleaned by the organization prior to sterilization and/or its use
Since you are manufacturing non sterile accessory, it is subject meet the clause 7.5.1.2.1 b) of ISO 13485 “product is supplied non-sterile to be subjected to a cleaning process prior to sterilization and/or its use”, As I said in my previous reply you need to have a minimum control in cleanliness of the product. You can use disinfectants to clean the accessory, prior to use check which type of disinfectants withstand your product (for eg. Iso prophyl alcohol). Also, do the validation study before and after cleaning with disinfectant. (This can be verify through bio-burden analysis and find out the percentage of microbial reduction level). Based on the result, you can have a control on the specific process.
Also, I suggesting that do the sterility validation in different ways that use cleaned material (packed condition), sterilize by ETO and do sterility test. (if don’t have in-house laboratory do the test in outside testing agencies)
If the test failed from the sterility, the one of the reason is high microbial load in your accessory/product.
Regards
S. Subramaniam