OK,....then the main question would be; inside ambulance, if everything accessible parts (chassis ambulance...rails for suporting medical devices, beds etc, or medical devices enclosure....) have to be treated like applied parts? (There are medical devices (for example; electronic oxygen conserver) that all accesible parts (for example...its enclosure) have to be treated like applied parts.Hi, Please note if your equipment is used in ambulance (EMS environment), its applied part should be type F (BF or CF). the type B applied part isn't acceptable. As you stated, type F applied design should revolve your problem, also this is mandatory.
Thanks