Hi there, we were using external calibration house to calibarte all ours furnace in accordance with AMS 2750 Rev E, we discovered that they were using on field equipment (T/C calibrator) to calibrate on field recorder and TUS thermocouples.During NADCAP audit we had NCR for the fact that on field calibrator wasn't calibrate on correct thermocouples types in this case type N.
what you can see on COC of field test equipment was done on type K thermocouple to cover range between 0-1100*C and on type N to cover range from 600-1000*C.but they used this equipment to calibrate TUS recorder on type N thermocouple on the range from 100 to 550*C and they did the same for TUS recorder(auditor base on this invalidated 10 surveys). Is it acceptable???we contacted calibration house and they replaied that it does not metter as calibartor was calibrate on 0-1100*C on type K and has been found to meet criteria so it must work on diffrent types to
any comments would be appreciate
what you can see on COC of field test equipment was done on type K thermocouple to cover range between 0-1100*C and on type N to cover range from 600-1000*C.but they used this equipment to calibrate TUS recorder on type N thermocouple on the range from 100 to 550*C and they did the same for TUS recorder(auditor base on this invalidated 10 surveys). Is it acceptable???we contacted calibration house and they replaied that it does not metter as calibartor was calibrate on 0-1100*C on type K and has been found to meet criteria so it must work on diffrent types to