Hello everybody,
We were checking IEC 60601 third edition rule and we found some points we have questions about. We will explain them one by one:
This rule (IEC 60601 third edition) stablishes the following limit in section c). It establishes a 500 µA limit for TOUCH CURRENT in SINGLE FAULT CONDITIONS and a 5 mA limit in section d) for EARTH LEAKAGE CURRENT in NORMAL CONDITIONS. I have attached on a screenshot the corresponding imagen below:
So, if we look at TOUCH CURRENT in SINGLE FAULT CONDITIONS test explanation, the definition is the one remarked on blue:
By other hand, if we see EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test explanation, the definition is:
So, our question is:
In a medical equipment class I with accessible conductive parts in the enclosure that are normally protectively earthed, are both tests equal but with different limits and if in EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test we are not under 500 µA it is impossible to meet the limit of TOUCH CURRENT in SINGLE FAULT CONDITIONS?
Or, it is that TOUCH CURRENT in SINGLE FAULT CONDITIONS must be applied on conductive parts of the enclosure that are not protected grounded?
Any help with this issue would be grateful.
Pads38
Moderator
As this has not yet been picked up I shall try and help...
The tests you enquire about are quite separate and different tests. Perhaps the clearest way to see this is to examine Figures 13 and 14 and notice the different use of the Measuring Device (MD).
In figure 13, for Earth leakage, the MD is in the earth conductor so measures all currents. Figure 14 it is shown ending in an arrow - this is meant to indicate that it is a measuring probe, which is then used all over the enclosure. For plastic parts or enclosure it is used in conjunction with metal foil.
You will only get a reading for touch current of protectively earthed metal parts in the Single Fault Condition of the earth connection open, effectively when they are no longer protectively earthed.
Alba Fernández
Hello!
First, thank you so much for your replay. But I still can´t understand completly the answer. I mean, according to Figure 14, in Single Fault Condition, if we messure with the MD between earth and all metalic parts that were connected to earth, but now not for the SFC, we are meassuring exaclty as in Figure 13 where all parts are earth connected.
The point is that the limit of touch current in Single Faut Conditions is well bellow the limit of earth leakage and MD is the same one and the test is also the same one so, we understand that, if Single Fault Conditions Touch Current limits are not met, it seems impossible to meet the test.
Thank you in advance. It is quite complex
Conaider
Hi Alba,
Fig. 14 in SFC with Protective Earth open would be the same as Fig 13. in NC if we measured Touch Current in an enclosure connected to Protective Earth.
However, a key point to consider is that you shall measure Touch Current for parts that are not protectively earthed. You don't measure TC on earthed enclosures.
I hope this helps to clarify.
Alba Fernández
Hi everybody,
First, thank you so much for your answers, there are being so usefull!
Regarding to the last replay, I think there is something that I am missing because, if you look at one of the first images I posted (I attach it in this post too), concretly imagen with point 8.7.4.6, you will see remarked in blue a quote when it is said that TC can be messured in a Protectively Earthed part in SFC.
So, in this case, Fig. 14 ans Fig 13. would be the same.
Pads38
Moderator
Touch Current can be Measured in a part that is Protectively Earthed - but only in the SFC of Earth Open.
Then it is, effectively, a conductive part that is NOT Protectively Earthed.
Alba Fernández
Thank you so much for your answers. So, after your replay, I understand this confirm what we said that is:
if in EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test we are not under 500 µA it is impossible to meet the limit of TOUCH CURRENT in SINGLE FAULT CONDITIONS in a part that is Protectively Earthed but in the SFC of Earth Open this conductive part is NOT Protectively Earthed.
Right?
Kuldeep Singh
Hi everyone,
We are manufacture of X-Ray medical products and perform different tests as per IEC 60601-1 in our internal lab.
We have Leakage current tester (Hioki Make) for single phase machine and able to perform leakage current tests on single phase machine easily. But we face difficulties in three phase machine. Is it applicable to three phase machine ? because it is a difficult task to create single fault condition in three phase machine.
Is there any Leakage current tester for three phase ?
EdinLipka
Hi....Where a solitary shortcoming condition isn't applied, the hardware is supposed to be in "typical condition" (NC). Notwithstanding, it is essential to comprehend that even in this condition, the presentation of specific tests may bargain the methods for assurance against electric stun. For instance, if earth spillage flow is estimated in ordinary condition, the impedance of the estimating gadget in arrangement with the defensive earth conveyor implies that there is no viable advantageous security against electric stun.
K Alternatives to Metal Foil for touch current measurement of enclosures made of insulating material. IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C Touch Current for Exposed Charging Contacts on Bodyworn Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 SIP/SIO Isolation and Touch current requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-2-49 - Touch Current (Enclosure Leakage Current) requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Touch Current Measurement and Limits of 42.4V peak or 60Vdc IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R Should we scrape off the enamel when measure the touch current? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Are Adapter Plugs Tested for Touch Current IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W 'Touch Current' test - Seeking clarity about clause 8.7.4.6 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B IEC 61010, Touch currents related to burns at higher frequencies (cl. 6.3, A.3) Other Medical Device Related Standards 0
B Accessible parts per 8.4.2 b) - Touch currents in combination with MOOP or MOPP? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 6
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
A IEC 60601-1 Impact test on glass touch controller IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Touch Protected Plugs used for Peripheral Nerve Stimulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Hershal Now for a touch of lunacy Coffee Break and Water Cooler Discussions 0
R ESD issue with capacitive touch screen IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M Zeiss U-Soft versus Uni-Touch CMM Software Calibration and Metrology Software and Hardware 7
Wes Bucey Touch Pad on Laptop - Make Me Smart After Work and Weekend Discussion Topics 6
M California Proposition 65 Compliant (no more than 600 ppm touch area) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Power switch requirement and on/off touch panel marking IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Marc An iMac "Touch" in the future? After Work and Weekend Discussion Topics 2
R Starrett AV612 (combo vision and touch probe) Galileo Instrument Software Validation General Measurement Device and Calibration Topics 3
T Touch Screen Capable Software suggestions for SPC (Statistical Process Control) Quality Assurance and Compliance Software Tools and Solutions 4
9 Another Touch on Management Commitment - Open Corrective Actions Quality Manager and Management Related Issues 6
Sidney Vianna Microsoft Surface techno gadget - Multi-Touch Table After Work and Weekend Discussion Topics 3
C APQP Craftsmanship requirements - What customers see, touch, hear, smell, & how used APQP and PPAP 5
Forum Administrator Operator Instructions and Lot Control - Touch Screens? Document Control Systems, Procedures, Forms and Templates 1
H 60601 - Patient leakage current via the patient connection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Using current gauge to verify masters General Measurement Device and Calibration Topics 1
R Help Understanding Current Role; Distributor, Manufacturer, or Importer EU Medical Device Regulations 4
C IEC60601-1: Impedance and current-carrying capability (8.6.4) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Leakage current measurement without isolating transformer IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F IEC 60601-2-54:2009 Accuracy of CURRENT TIME PRODUCT IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Precision Current Loop for Current Probe Calibration Reliability Analysis - Predictions, Testing and Standards 2
N Auxiliary current limit with coil cell powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B Leakage Current AC-DC discrimination per cl. 8.7.3 b) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Earth (ground) leakage current routine test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T How can I determine the current level of exposure to talc in the workplace? Occupational Health & Safety Management Standards 6
H Prodigit DC Electronic Load - Drawing current but showed 0 on the screeen Manufacturing and Related Processes 0
D Add new device to current license Canada Medical Device Regulations 3
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W FMEA - Current control and occurrence rating FMEA and Control Plans 3
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 5
W Need for current design or process control FMEA and Control Plans 2
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Best way to keep up with pharmaceutical quality knowledge and current affairs Service Industry Specific Topics 1
T Iec 60601 Impedance and current-carrying capability IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Informational TGA – Current status of breast implant products in Australia Medical Device and FDA Regulations and Standards News 0
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1