Hello everybody,
We were checking IEC 60601 third edition rule and we found some points we have questions about. We will explain them one by one:
This rule (IEC 60601 third edition) stablishes the following limit in section c). It establishes a 500 µA limit for TOUCH CURRENT in SINGLE FAULT CONDITIONS and a 5 mA limit in section d) for EARTH LEAKAGE CURRENT in NORMAL CONDITIONS. I have attached on a screenshot the corresponding imagen below:
So, if we look at TOUCH CURRENT in SINGLE FAULT CONDITIONS test explanation, the definition is the one remarked on blue:
By other hand, if we see EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test explanation, the definition is:
So, our question is:
In a medical equipment class I with accessible conductive parts in the enclosure that are normally protectively earthed, are both tests equal but with different limits and if in EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test we are not under 500 µA it is impossible to meet the limit of TOUCH CURRENT in SINGLE FAULT CONDITIONS?
Or, it is that TOUCH CURRENT in SINGLE FAULT CONDITIONS must be applied on conductive parts of the enclosure that are not protected grounded?
Any help with this issue would be grateful.
We were checking IEC 60601 third edition rule and we found some points we have questions about. We will explain them one by one:
This rule (IEC 60601 third edition) stablishes the following limit in section c). It establishes a 500 µA limit for TOUCH CURRENT in SINGLE FAULT CONDITIONS and a 5 mA limit in section d) for EARTH LEAKAGE CURRENT in NORMAL CONDITIONS. I have attached on a screenshot the corresponding imagen below:

So, if we look at TOUCH CURRENT in SINGLE FAULT CONDITIONS test explanation, the definition is the one remarked on blue:

By other hand, if we see EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test explanation, the definition is:

So, our question is:
In a medical equipment class I with accessible conductive parts in the enclosure that are normally protectively earthed, are both tests equal but with different limits and if in EARTH LEAKAGE CURRENT in NORMAL CONDITIONS test we are not under 500 µA it is impossible to meet the limit of TOUCH CURRENT in SINGLE FAULT CONDITIONS?
Or, it is that TOUCH CURRENT in SINGLE FAULT CONDITIONS must be applied on conductive parts of the enclosure that are not protected grounded?
Any help with this issue would be grateful.