TS16949 says - records of calibration/verificatin must - then it lists what the records must have.
But anytime you state - you "verify" gages at specified intervals in your procedures or tell the auditor you do, do you have to show records of this verification?
Thank you
kat.k.
I believe there may be a little more to this question than meets the eye. What if you have a micrometer on an
annual calibration cycle, but you verify it
daily before use? Are you required to maintain this full list of information for both the annual calibration and each day's verification?
(These are not listed in the same order as TS 16949)
1) equipment identification, including the measurement standard against which the equipment is calibrated
2) statements of conformity to specification after calibration/verification
3) any out-of-specification readings as received for calibration/verification
4) revisions following engineering changes
5) an assessment of the impact of out-of-specification condition
6) notification to the customer if suspect product or material has been shipped
I would suggest that items 1 & 2 need to be included in all records, including both the annual calibration and the daily verification. The annual calibrations would probably include the actual as-found/as-left readings (numerical data) and the standards used for calibration. The daily verifications could be as simple as pass/fail and the standard(s) against which they were verified. Items 3-6 would only be needed in applicable situations such as when the gage is found out-of-tolerance, or when modifications to gaging are made.