Hoping someone can help...
Here are the facts:
We make mechanical (non electrical) Class ! devices. These attach to surgical tables and are used to position patients in the operating room.
When we design our products, we do overload testing to a 2.2 factor of safety. We base this testing on a test found in:IEC 60601-2-46Medical electrical equipment –Part 2-46articular requirements for the safety of operating tables, though we do not reference the standard in our CE self certification
Life testing is performed based on 5 years of use within the range of the specification (1X) rather than the 2.2 overload weight.
Here are the questions:
From a regulatory standpoint, is there a requirement (or if not a requirement, the expectation) that the 2.2 safety factor must be maintained through out the life of the product, or only that the product remain within specification.? The 60601 standard does not provide any time period for 2.2 load testing.
Assuming we don't use the 2.2 factor of safety after initial load testing, would the expectation be that the 2,2 should be maintained at least when FAI tested out of box by a customer, or should the testing be done to the spec?
Thanks much for any thoughts on this!
Here are the facts:
We make mechanical (non electrical) Class ! devices. These attach to surgical tables and are used to position patients in the operating room.
When we design our products, we do overload testing to a 2.2 factor of safety. We base this testing on a test found in:IEC 60601-2-46Medical electrical equipment –Part 2-46articular requirements for the safety of operating tables, though we do not reference the standard in our CE self certification
Life testing is performed based on 5 years of use within the range of the specification (1X) rather than the 2.2 overload weight.
Here are the questions:
From a regulatory standpoint, is there a requirement (or if not a requirement, the expectation) that the 2.2 safety factor must be maintained through out the life of the product, or only that the product remain within specification.? The 60601 standard does not provide any time period for 2.2 load testing.
Assuming we don't use the 2.2 factor of safety after initial load testing, would the expectation be that the 2,2 should be maintained at least when FAI tested out of box by a customer, or should the testing be done to the spec?
Thanks much for any thoughts on this!